upgrade to MH 11.1

Hello,

i upgraded my MH QUANT 10.1 to 11.1 and since then the batch.script doesn't work for me, we are using a modified component.config, do you know where the error could be? This component.config worked with all 10 versions of masshunter (10,10.1 and 10.2)

Thank you.

Parents
  • First of all, thank you for posting publicly, I had not noticed this interface feature of MassHunter Quant, it looks very useful.

    Were you able to get it working in 11.1?

    If you're allowed to share your "IScriptable" folder, I would be very interested in seeing an use example. I am University/Government employed in Denmark, if that helps regarding permissions. Slight smile

    I am aware it will not function fully on my machine, but csc.exe indicates some compiling failing at the initialization, and you don't seem to be referencing any internal dll, so I should get far enough that I might be able to reproduce or not, and if it doesn't fail on my machine, you can get a hint at what is wrong.

  • Hi, sorry i can't share our script, but i can share testing script for compilation purposes. It still doesn't work on MH 11.1, compilation of methods from IScriptableInterface does not work when uploading a batch. You must create IScriptable folder like this C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt"> and copy that batch_script.txt into , and copy component.config.xml here C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin, make backup of that component.config.

    <?xml version="1.0" encoding="utf-8"?>
    <components>
      <defaultnamespace>Agilent.MassSpectrometry.DataAnalysis.Quantitative</defaultnamespace>
      <references>
        <reference framework="true" assembly="System.Data.dll"/>
        <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
        <reference framework="true" assembly="System.Xml.dll"/>
        <reference framework="true" assembly="System.Web.Services.dll"/>
        <reference framework="false" assembly="Quantitation.dll"/>
        <reference framework="false" assembly="QuantUtils.dll"/>
        <reference framework="false" assembly="ScriptableInterfaces.dll"/>
        <reference framework="true" assembly="System.Linq.dll"/>
        <reference framework="true" assembly="System.Core.dll"/>
        <reference framework="true" assembly="System.Xml.Linq.dll"/>
        <reference framework="true" assembly="System.Windows.Forms.dll"/>
        <reference framework="true" assembly="System.Threading.Tasks.dll"/>
        <reference framework="true" assembly="System.Drawing.dll"/>
        <reference framework="true" assembly="System.Net.dll"/>
        <reference framework="true" assembly="System.IO.dll"/>
      </references>
      <interface name="IScriptableSample">
        <implementation assembly="CompoundMath_Script.dll">
          <script name="CompoundMath_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="ICustomExpressions">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.cs">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <!--<interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>-->
      <interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
              <reference framework="true" assembly="System.Data.dll"/>
              <reference framework="true" assembly="System.Linq.dll"/>
              <!--<reference framework="true" assembly="System.Linq.Enumerable.dll"/>-->
              <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
              <reference framework="true" assembly="System.Collections.dll"/>
              <reference framework="true" assembly="System.Xml.Linq.dll"/>
              <reference framework="true" assembly="System.IO.dll"/>
              <reference framework="true" assembly="System.Xml.dll"/>
              <reference framework="true" assembly="System.Core.dll"/>
              <reference framework="true" assembly="System.Threading.Tasks.dll"/>
              <reference framework="true" assembly="System.Net.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableOutlier" instanceName="Accuracy">
        <implementation assembly="OutlierAccuracy_Script.dll">
          <script name="OutlierAccuracy_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableSample">
        <implementation assembly="Sample_Script.dll">
          <script name="Sample_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableCompound">
        <implementation assembly="Compound_Script.dll">
          <script name="Compound_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQuantifierIon">
        <implementation assembly="QuantifierIon_Script.dll">
          <script name="QuantifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQualifierIon">
        <implementation assembly="QualifierIon_Script.dll">
          <script name="QualifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
    </components>
    using System;
    using Agilent.MassSpectrometry.DataAnalysis.Quantitative;
    
    namespace ScriptTest
    {
        internal class Batch_Script
        {
        }
    
        public class ScriptableBatch : IScriptableBatch
        {
            public void _OnBatchAnalysisStarting(QuantitationDataSet dataSet, short batchId)
            {
                // ALS Implementation
            }
    
            public void _OnBatchAnalyzed(QuantitationDataSet dataSet, short batchId)
            {
                // ALS Implementation
            }
    
            public void _OnBatchLoaded(QuantitationDataSet dataSet, short batchId)
            {
                // ALS Implementation
            }
        }
    }

Reply
  • Hi, sorry i can't share our script, but i can share testing script for compilation purposes. It still doesn't work on MH 11.1, compilation of methods from IScriptableInterface does not work when uploading a batch. You must create IScriptable folder like this C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt"> and copy that batch_script.txt into , and copy component.config.xml here C:\Program Files\Agilent\MassHunter\Workstation\Quant\bin, make backup of that component.config.

    <?xml version="1.0" encoding="utf-8"?>
    <components>
      <defaultnamespace>Agilent.MassSpectrometry.DataAnalysis.Quantitative</defaultnamespace>
      <references>
        <reference framework="true" assembly="System.Data.dll"/>
        <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
        <reference framework="true" assembly="System.Xml.dll"/>
        <reference framework="true" assembly="System.Web.Services.dll"/>
        <reference framework="false" assembly="Quantitation.dll"/>
        <reference framework="false" assembly="QuantUtils.dll"/>
        <reference framework="false" assembly="ScriptableInterfaces.dll"/>
        <reference framework="true" assembly="System.Linq.dll"/>
        <reference framework="true" assembly="System.Core.dll"/>
        <reference framework="true" assembly="System.Xml.Linq.dll"/>
        <reference framework="true" assembly="System.Windows.Forms.dll"/>
        <reference framework="true" assembly="System.Threading.Tasks.dll"/>
        <reference framework="true" assembly="System.Drawing.dll"/>
        <reference framework="true" assembly="System.Net.dll"/>
        <reference framework="true" assembly="System.IO.dll"/>
      </references>
      <interface name="IScriptableSample">
        <implementation assembly="CompoundMath_Script.dll">
          <script name="CompoundMath_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="ICustomExpressions">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\CustomExpressions_Script.cs">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <!--<interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
            </references>
          </script>
        </implementation>
      </interface>-->
      <interface name="IScriptableBatch">
        <implementation assembly="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.dll">
          <script name="C:\MassHunter\Scripts\Quant\IScriptable\Batch_Script.txt">
            <references>
              <reference framework="false" assembly="BatchAnalysis.dll"/>
              <reference framework="true" assembly="System.Data.dll"/>
              <reference framework="true" assembly="System.Linq.dll"/>
              <!--<reference framework="true" assembly="System.Linq.Enumerable.dll"/>-->
              <reference framework="true" assembly="System.Data.DataSetExtensions.dll"/>
              <reference framework="true" assembly="System.Collections.dll"/>
              <reference framework="true" assembly="System.Xml.Linq.dll"/>
              <reference framework="true" assembly="System.IO.dll"/>
              <reference framework="true" assembly="System.Xml.dll"/>
              <reference framework="true" assembly="System.Core.dll"/>
              <reference framework="true" assembly="System.Threading.Tasks.dll"/>
              <reference framework="true" assembly="System.Net.dll"/>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableOutlier" instanceName="Accuracy">
        <implementation assembly="OutlierAccuracy_Script.dll">
          <script name="OutlierAccuracy_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableSample">
        <implementation assembly="Sample_Script.dll">
          <script name="Sample_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableCompound">
        <implementation assembly="Compound_Script.dll">
          <script name="Compound_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQuantifierIon">
        <implementation assembly="QuantifierIon_Script.dll">
          <script name="QuantifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
      <interface name="IScriptableQualifierIon">
        <implementation assembly="QualifierIon_Script.dll">
          <script name="QualifierIon_Script.txt">
            <references>
            </references>
          </script>
        </implementation>
      </interface>
    </components>
    using System;
    using Agilent.MassSpectrometry.DataAnalysis.Quantitative;
    
    namespace ScriptTest
    {
        internal class Batch_Script
        {
        }
    
        public class ScriptableBatch : IScriptableBatch
        {
            public void _OnBatchAnalysisStarting(QuantitationDataSet dataSet, short batchId)
            {
                // ALS Implementation
            }
    
            public void _OnBatchAnalyzed(QuantitationDataSet dataSet, short batchId)
            {
                // ALS Implementation
            }
    
            public void _OnBatchLoaded(QuantitationDataSet dataSet, short batchId)
            {
                // ALS Implementation
            }
        }
    }