Upgrading from Windows 7 to Windows 10. Is Masshunter and Chemstation compatible?

Hello,

 

Our lab is planning to upgrade several of our workstation PC's from Windows 7 to Windows 10. I would like to know if software would be compatible with Windows 10.

 

Here is a list of the Agilent Programs:

 

MassHunter B.07.04.2260
MassHunter B.07.00 SP2.1654
MassHunter 4.4 workstation version C.01.04
OpenLab CDS C.01.07[27]
OpenLab CDS C.01.09[150]

 

Will we also need to product keys for the re-installation? I am worried we will be unable to locate the physical copies of the product keys.

Parents
  • Hello,

     

    I can only speak definitely for your OpenLab ChemStation software.

     

    OpenLab CDS ChemStation C.01.07 [27] is NOT supported on Windows 10. If you wish to run C.01.07 on Windows 10 you would need to upgrade the software to C.01.07 SR3 or SR4. 

     

    OpenLab CDS ChemStation C.01.09 is supported on Windows 10. 

     

    For MassHunter Acquisition for GCMS B.07.04 and Qual B.07.00 SP2, I believe these are only supported on Windows 7 SP1 and would need to be upgraded if you are planning on upgrading your operating system. 

Reply
  • Hello,

     

    I can only speak definitely for your OpenLab ChemStation software.

     

    OpenLab CDS ChemStation C.01.07 [27] is NOT supported on Windows 10. If you wish to run C.01.07 on Windows 10 you would need to upgrade the software to C.01.07 SR3 or SR4. 

     

    OpenLab CDS ChemStation C.01.09 is supported on Windows 10. 

     

    For MassHunter Acquisition for GCMS B.07.04 and Qual B.07.00 SP2, I believe these are only supported on Windows 7 SP1 and would need to be upgraded if you are planning on upgrading your operating system. 