Upgrading from Windows 7 to Windows 10. Is Masshunter and Chemstation compatible?

Hello,

 

Our lab is planning to upgrade several of our workstation PC's from Windows 7 to Windows 10. I would like to know if software would be compatible with Windows 10.

 

Here is a list of the Agilent Programs:

 

MassHunter B.07.04.2260
MassHunter B.07.00 SP2.1654
MassHunter 4.4 workstation version C.01.04
OpenLab CDS C.01.07[27]
OpenLab CDS C.01.09[150]

 

Will we also need to product keys for the re-installation? I am worried we will be unable to locate the physical copies of the product keys.

Parents Reply Children
  • Hello,

     

    There should not be any problem upgrading to C.01.09. Just make sure that you have enough v1.9 licenses in subscribenet if you do plan on upgrading. 

     

    I am also going to note that in-place OS upgrades are not supported. My recommendation for upgrading to Win10 would be to to just use a fresh Win10 image and then install OpenLab ChemStation on the new image. Make sure to backup all of your methods, data, etc... before hand so that they can be transferred over to the new Win10 install. 