We have the 6470 spectrometer that was turned off to do electrical work in the building. When we turned it back on, the communication is not taking place between the QQQ and the software. we restarted the system but the problem still persists
We have the 6470 spectrometer that was turned off to do electrical work in the building. When we turned it back on, the communication is not taking place between the QQQ and the software. we restarted the system but the problem still persists
Hi
Can you ping the 6470? Open a command prompt, then ping 192.168.254.12.
Hi,
thank you for your reply
we pinged the 6470 with 192.168.254.12 and also 192.168.254.10 and 192.168.254.11
but all these manipulations did not work
Hi,
Please check instrument computer ethernet ipv4 adress.. It should be 192.168.254.10, submask 255 255 255 0 . After that, try ping computer and all system.
Hi,
Thank you for your interest
We did all these operations but it didn't work
Can you hear "beep" in ~10 sec after switching the MS on? Can you hear diverter valve switching? If not, maybe the Linux inside the MS won't load.
So, your LC 192.168.254.11 also doesn't ping? If you can't ping anything then maybe your network switch doesn't work.
Hi,
1/ Can you hear "beep" in ~10 sec after switching the MS on? Yes
2/ Can you hear diverter valve switching? Yes
Thank you
We don't understand why it doesn't ping any of these addresses. We simply turned off the LCMSMS following the procedure
We don't understand why it doesn't ping any of these addresses. We simply turned off the LCMSMS following the procedure