Hi,
What is the acceptable criteria for the Air and Water leak check on a GC-MS?
For 69, 18, and 28, I have a relative abundance of 100, 1.61, and 1.15. This seems really low. Can you please give me suggestions as to what I should do?
Thank you!
Hi,
What is the acceptable criteria for the Air and Water leak check on a GC-MS?
For 69, 18, and 28, I have a relative abundance of 100, 1.61, and 1.15. This seems really low. Can you please give me suggestions as to what I should do?
Thank you!
Hi
Check above link to get answer
Thank you, ISH. It seems that my abundances are low for 69, 219, and 502 m/z. I have 231360, 57248, and 2577, for 69, 219, and 502, respectively. Do you have any suggestions? Thank you.
Thank you. I am wondering if the problem is associated with air in the column/conditioning the column. The Hi-Vac also looks to be the problem.
Please attached autotune report..
Can you see it now?
When source of MS was cleaned then that time any GC maintenance performed means consumables like septa/ liner/ gold seal etc were replaced or not
Not those consumables, but consumables in the source were replaced.
Share autotune report like this...
I am unable to get that autotune report due to the age of my software. We do not have a PDF software on the computer and the computer is not connected to the internet
See, water can be removed by baking of ms and conditioning of column but for air leak need to check all points mentioned in documents i shared..
Check septa nut is tight properly or not..better put new septa..
Autotune print out is coming or not.
No, Autotune print won't work either. The only thing I can generate is what I uploaded earlier
Ok...change septa then check.