Hi,
What is the acceptable criteria for the Air and Water leak check on a GC-MS?
For 69, 18, and 28, I have a relative abundance of 100, 1.61, and 1.15. This seems really low. Can you please give me suggestions as to what I should do?
Thank you!
Hi,
What is the acceptable criteria for the Air and Water leak check on a GC-MS?
For 69, 18, and 28, I have a relative abundance of 100, 1.61, and 1.15. This seems really low. Can you please give me suggestions as to what I should do?
Thank you!
If you can share auto tune and tune evaluation report then i can say anything..
Meanwhile, you can check level of PFTBA.
Last time, when source was cleaned.
How is purity of carrier? Sometimes, moiture from carrier or from saturated gas filter( if connected)...
Source was last cleaned on Sept 8 by an Agilent service engineer. Column has since been change. Do you think there is air in the column? Purity of carrier is 99.99% helium.
Tune report should be at the following link. I have an older computer that doesn't have a PDF software.
Hi
I did not get autotune report but you can go through below link to cross check ur autotune report.
community.agilent.com/.../reviewing-the-autotune-report-for-agilent-5977-gc-msd-systems-in-ei-mode