Hi,
What is the acceptable criteria for the Air and Water leak check on a GC-MS?
For 69, 18, and 28, I have a relative abundance of 100, 1.61, and 1.15. This seems really low. Can you please give me suggestions as to what I should do?
Thank you!
Hi,
What is the acceptable criteria for the Air and Water leak check on a GC-MS?
For 69, 18, and 28, I have a relative abundance of 100, 1.61, and 1.15. This seems really low. Can you please give me suggestions as to what I should do?
Thank you!
Hi
Check above link to get answer
Thank you, ISH. It seems that my abundances are low for 69, 219, and 502 m/z. I have 231360, 57248, and 2577, for 69, 219, and 502, respectively. Do you have any suggestions? Thank you.
Not those consumables, but consumables in the source were replaced.
Share autotune report like this...
I am unable to get that autotune report due to the age of my software. We do not have a PDF software on the computer and the computer is not connected to the internet
See, water can be removed by baking of ms and conditioning of column but for air leak need to check all points mentioned in documents i shared..
Check septa nut is tight properly or not..better put new septa..
Autotune print out is coming or not.
Autotune print out is coming or not.
No, Autotune print won't work either. The only thing I can generate is what I uploaded earlier
Ok...change septa then check.
And check this link to points where leak possible in ms..
I changed the septa, ran the tune report, and this is what I got.
Here EM volt is showing 171...Am i right..