GCMS tuning stopped 5975C

Hello All,

I'm seeing an issue that I haven't come across before - have just changed the column and pumped back down yesterday. Came to do an air and water check this morning and i get the below: 

Tune action stopped:-222{SP00}:MSC:RDY ON Parameter ento value 64.256 out of range [0.000, 64.000]

I didn't do anything with the source yesterday, just changed the column....

Thanks in advance!

Parents Reply
  • yes vent valve was closed properly - Just to update, today (after leaving the system to settle over the weekend) I have run a Tune and Tune evaluation and it has passed. There are still a high number of peaks in scan section so indicating source could do with a clean but tune portion is now passing. I'm just running a blank to see what the chromatography looks like. 