GCMS tuning stopped 5975C

Hello All,

I'm seeing an issue that I haven't come across before - have just changed the column and pumped back down yesterday. Came to do an air and water check this morning and i get the below: 

Tune action stopped:-222{SP00}:MSC:RDY ON Parameter ento value 64.256 out of range [0.000, 64.000]

I didn't do anything with the source yesterday, just changed the column....

Thanks in advance!

Parents
  • Hi ISH, yes all was fine with column installation and pump down was efficient. The issue seems to be with the entrance lense offset - I loaded a previous tune and managed to run an autotune however it is failing due to too high 503/502 ratio. I've seen some advice that this could be due to HED needing cleaning or a dirty source perhaps? The Ion focus was consistently 90.2 however in the latest tune it is now 71.2 so something is not right this week from last week and previous tunes!

    Tune last week had 503/502 ratio of 10.13 where as today it is 13.28. The peak shape of the 502 is also not as neat, slight splitting in the 503 peak and noisier baseline either side.

    If anyone has any advice it would be welcome!

  • I tried to add the tune reports as pdfs but couldn't upload.... Thanks both ISH and paulsalverda for your replies, we got the system second hand so I don't have a 'local' agilent engineer I can call on! I have no experience of removing and cleaning the source of this system but it is looking likely I'll need to. 

    The curious thing is that the only difference between the good tune a week prior and the bad tune last week is that I had to vent and clear some debris out of the analyser chamber...I didn't actually touch the source or run any samples so its strange that suddenly the tune has an issue :(  

  • Hi

    Just take photo of tune report then you can upload easily..

Reply Children
Was this helpful?