FID drift/noise report with Openlab CDS 2.7

I am searching a way to create a report for the FID drift and noise with Openlab CDS 2.7. 

Is it possible? 

Thank you 

Parents Reply Children
  • Hi, you dant have a chance, in OL2 you can choose only 1 period, maybe looooooong way, create a 10 different processing methods , generete 10 results and then create cross-sequence report 

    Or,  maybe easiest way -  choose in system suitability choose 10 minutes period for noise determination and calculate , you will have the average value

  • Ok. And there is no way we can have the drift value over a certain period of time with openlab software? 

  • Hello,

    You can get the drift for a specific time period as shown below in the first image. You could get the P2P drift corrected noise for 10 1 min segments in the second image, note I am only doing 5 in the example. This will require processing with 2 methods as you cannot mix noise calculations. You could manually integrate 10 peaks in the second example, but if you have to do this more than once use the EZChrom integrator. Using that integrator you can setup 10 peaks 1 min apart in the range where you want to calculate the noise using the manual peak event. Note, in the example I am using the drift corrected P2P noise. 

    Marty Adams 