This Information Applies To: Agilent 6820, 6890, 7820A GC systems with a Split/Splitless (SSL) inlet
Issue
Non-volatile sample matrix components will accumulate in the inlet liner and impact on analytical performance. The inlet liner should be changed regularly as part of your routine maintenance...
This Information Applies To: Agilent headspace samplers (HS)
Issue
GC cycle time needs to be calculated correctly. Too short a GC cycle time will reduce sample quality and possibly stop the analysis prematurely. Too long GC cycle time can cause reduced sample throughput as vials wait longer than...
This Information Applies To: Agilent 7820A GC
Issue
7820A GC is designed with simplified keypad that comes with five buttons. They are Stop, Prep Run , Start, Up and Down arrow buttons. Changing the IP address requires a different method compared to instruments with Keypad or touch screen.
...
This Information Applies To: 5977A and 5977B
Content: This procedure is to guide the installation of a capillary column on 5977A/B GC/MS with a standard column nut or self-tightening nut.
Warning : a. The GC and MS operate at high temperatures. Do not touch any parts of GC and MS until...
This Information Applies To: Agilent 7697A headspace sampler
Issue
During sequence running, there is error message: "Sequence aborted: Data System Not Ready" displayed on the 7697A display screen.
Resolution
Check GC cycle time. If the GC cycle time is set too short in the GC Method the...