This Information Applies To: Agilent 7693A Automatic Liquid Sampler
Issue:
The automatic liquid sampler may need to be replaced, in case it is worn or dirty.
Steps to follow
To remove the syringe:
Open the injector door.
Slide the syringe carriage to the top position if necessary...
This information applies to: Agilent Intuvo 9000, 8860, 8890, 7890A/B, 7820A, 6890, 6850 GC systems with Split/Splitless, or multimode inlets.
Issue:
This article describes how to replace the inlet septum for Split/Splitless and multimode inlets for all Agilent GC systems. Regular replacement of...
This information applies to: Agilent 8890 GC system.
Issue
Turn off the GC safely. This article outlines both short term (one week or less) and long term (more than one week) shutdown of the GC.
Requirements
Warning : Do not touch the power source at the back of the GC if the cable...
This information applies to: Agilent 8890 GC system.
Issue
Turning on the 8800 GC safely.
Requirements
A dedicated power line to the GC is required.
Caution : Users need to make sure the oven voltage requirements match the power socket voltage, otherwise they will run the risk...
This information applies to: Agilent GC/MS systems running MassHunter data acquisition software for GC/MS (B.07.x - B.10.x).
Issue
Set up a GC only (without the mass spectrometer) configuration in MassHunter.
Steps to Follow
Close any acquisition instances.
Open the Agilent GC/MS instrument...