Hi,
I have recently seen the "Carousel Error 8" on one of our 7697A Headspace samplers and was curious if anyone knows what this particular error code indicates is the issue?
I has happened twice in my last run, about every 5 or so vials.
Thanks!
Hi,
I have recently seen the "Carousel Error 8" on one of our 7697A Headspace samplers and was curious if anyone knows what this particular error code indicates is the issue?
I has happened twice in my last run, about every 5 or so vials.
Thanks!
Need to update the headspace firmware to A.01.08.3
Thanks! Where do we get the firmware update? Can we download from Agilent and update ourselves?
Hello chrom . You can download the firmware utility from here: GC Firmware Update | Agilent. The procedure to update the firmware can be found in the following resources in the Agilent Knowledge Portal: Agilent GC Firmware Update Procedure - GC Articles - Gas Chromatography Portal - Agilent Community and Agilent GC Firmware Update - GC Videos - Gas Chromatography Portal - Agilent Community.
The resources show the procedure for GCs, but it is the same tool for the Agilent 7697A headspace sampler
Hello chrom . You can download the firmware utility from here: GC Firmware Update | Agilent. The procedure to update the firmware can be found in the following resources in the Agilent Knowledge Portal: Agilent GC Firmware Update Procedure - GC Articles - Gas Chromatography Portal - Agilent Community and Agilent GC Firmware Update - GC Videos - Gas Chromatography Portal - Agilent Community.
The resources show the procedure for GCs, but it is the same tool for the Agilent 7697A headspace sampler
Thanks!