Hi,
I have recently seen the "Carousel Error 8" on one of our 7697A Headspace samplers and was curious if anyone knows what this particular error code indicates is the issue?
I has happened twice in my last run, about every 5 or so vials.
Thanks!
Hi,
I have recently seen the "Carousel Error 8" on one of our 7697A Headspace samplers and was curious if anyone knows what this particular error code indicates is the issue?
I has happened twice in my last run, about every 5 or so vials.
Thanks!
Need to update the headspace firmware to A.01.08.3
Thanks! Where do we get the firmware update? Can we download from Agilent and update ourselves?
Thanks! Where do we get the firmware update? Can we download from Agilent and update ourselves?
Hello chrom . You can download the firmware utility from here: GC Firmware Update | Agilent. The procedure to update the firmware can be found in the following resources in the Agilent Knowledge Portal: Agilent GC Firmware Update Procedure - GC Articles - Gas Chromatography Portal - Agilent Community and Agilent GC Firmware Update - GC Videos - Gas Chromatography Portal - Agilent Community.
The resources show the procedure for GCs, but it is the same tool for the Agilent 7697A headspace sampler
Thanks!