Help getting OpenLabs CDS open - freezing on startup macros

When I try to open OpenLabs CDS (online version), it is freezing on "Initializing: Startup Macros".  I have restarted the GC and the computer multiple times, no change.  It is an Agilent 6890 with 7683B injector.  Any tips are appreciated!

Top Replies

Parents
  • Hello,

     

    As recommended, resetting the instrument's config files is a good place to start. On top of the config.reg file, I would also recommend resetting all of the files outlined in the following link (in C.01.09 these files are found in C:\ProgramData\Agilent Technologies\ChemStation\#\ where # is the instrument number):Autostart macro failed error 999 in OpenLAB CDS ChemStation 

     

    If resetting the config files does not resolve the problem, the next thing to check would be if you are specifying any custom macros to be initialized upon startup. Check in C:\Program Files (x86)\Agilent Technologies\ChemStation\CORE to see if there is a user.mac file present. If so, try renaming this file to user.old and seeing if you can then launch ChemStation. If you can, then it would indicate that a macro or command specified in the user.mac file is running into problems. 

  • Thanks and .

    I renamed the 4 files (config.reg, ChemStationPerInstrumentSettings.config, instrument.config, RapidControl.InstrumentConfig.xml). I do not have a user.mac file in the CORE folder.

    The program opened once and I reconnected to the GC.  However, when I try to close and reopen the program, I am now seeing a new error "Launching failed.  Value cannot be null. Parameter name: Command line argument of ChemMain process is empty".  The program freezes as soon as the window opens, before reaching the Startup Macros dialog.  Any ideas?

     

    I have also noticed that whenever I try to restart my computer, it says that process "CAG Server Executable (32-bit)" is preventing restart.

  • Hello,

     

    I would recommend that you run the Software Verification Tool (available under Start>Agilent Technologies) to qualify your installation. Make sure that all of the SV Reports that are generate come back with an overall status of Pass.

  • Hi,

    All of the SV reports have an overall status of pass and did not find any

    issues.

Reply Children
Was this helpful?