Printing Problem SpektrAA 5.3
Find maintenance, best practice, and troubleshooting articles.
Printing Problem SpektrAA 5.3
Hello
I have seen this before- if certain report settings are selected but no data has been collected (for example, you haven't completed a calibration but have selected calibration graph report to be included in the report settings) the report will not print. In your case, if you are trying to print a blank method, you will need to deselect almost everything that would require collected data in order to report in the report settings, and only include method parameters, notes, ect.
Hope that makes sense
Hello
I have seen this before- if certain report settings are selected but no data has been collected (for example, you haven't completed a calibration but have selected calibration graph report to be included in the report settings) the report will not print. In your case, if you are trying to print a blank method, you will need to deselect almost everything that would require collected data in order to report in the report settings, and only include method parameters, notes, ect.
Hope that makes sense
Many thanks for the solution. It works perfectly. Things can be so simple, if you know why.