Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
Greetings Agilent Community,
As with many of the errors I encounter with my system, these two happened after I foolishly left the lab to pursue a life outside the world of LCMS.
(1) From the instrument log, the first error read "Failed to run acquisition…
When attempting to submit a sequence in EZChrom, it fails to submit and returns this error message:
This happens each time a sequence is submitted. Any help would be appreciated!
Hi Agilent community.
I'm a LC-MS equipment user and I had a problem with the MassHunter Acquisition B 08.01 software.
When I try to open the Acquisition software, appears the software logo but immediately appears an error dialog box. This error dialog…
Hi everyone,
We have a 6546 QTOF, the following error keeps popping up during acquisition; Failed to start logbook server. A shared violation has occurred. How can I resolve this?. I'll appreciate your help. Thanks
This Information Applies To: Agilent 8890/8860 GC Systems
Issue
Early Maintenance Feedback is a useful feature on Agilent GC Systems and software that allows a user to monitor the maintenance needed on their GC. This feature allows for users to...
This Information Applies To: Agilent 8890/8860 GC Systems
Issue:
This procedure outlines the steps to view and clear the maintenance warnings on based on Agilent 8890/8860 GCs. You can also follow the onscreen instructions to perform maintenance...
This Information Applies To: Agilent 6000 Series LC/MS systems equipped with Agilent MS-40+ rotary vane pump.
Issue
Regular check of the rough pump fluid level in your MS40+ pump as part of regular 6000 LC/MS system maintenance checks.
Steps...
This Information Applies To the Following Products : All Agilent iFunnel and Agilent Jet Stream (AJS) 6400 Series LC/MS triple quadrupole systems, Agilent 6460A/C, 6470A, 6490A, and 6495A/B/C
Issue:
As LC Triple Quadrupole Mass Spectrometers (LC...