Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC/MS Portal
  • Agilent Knowledge Portal
LC/MS Portal
Articles Troubleshooting LC/TQ Contamination Issues
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC/MS Portal Home
  • LC/MS Portal Articles
  • LC/MS Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  21 Jan 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  LC/MS Systems
  • Component :  LC/MS Triple Quadrupole
  • Product Name :  6470 Triple Quadripole LC/MS 6475 Triple Quadripole LC/MS 6490 Triple Quadripole LC/MS 6495 Triple Quadripole LC/MS Ultivo Triple Quadrupole LC/MS
  • Task :  Troubleshooting
Related Articles
  • Installing an Ion Capillary in Triple Quadrupole LC/MS Systems Equipped With a Gate Valve

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve Issue Installing the Ion Transfer Capillary after cleaning or replacement in Agilent LC/MS instruments with a gate valve. Required Items Clean powder-free nitrile gloves LC/MS-grade isopropanol Caution : Ensure the source heated zones are cool before working on the instrument. Before installing the new ion capillary, the used ion capillary must be removed...
  • Removing the Ion Transfer Capillary from Triple Quadrupole LC/MS Systems Equipped With a Gate Valve

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems equipped with a gate valve. Issue Removing the Ion Transfer Capillary for maintenance on instruments with gate valve. The gate valve allows ion capillary removal and installation without having to vent the instrument. Required Items Clean powder-free nitrile gloves Capillary puller tool (supplied with your triple quadrupole LC/MS with a gate valve installed) Steps to follow Caution : Handling...
  • What to do if my LC-QQQ Autotune Fails?

    This Information Applies To: Agilent 6400 Series triple quadrupole LC/MS systems Issue Autotune fails to complete. Background Agilent 6400 LC/MS systems use an Autotune algorithm to optimize the sensitivity and mass resolution of the instrument. Sometimes when running the autotune procedure the autotune does not complete and the user is given an Autotune failed message in MassHunter. One of the main reasons for Autotune failure is a lack of sensitivity or abundance of the tuning ions...
  • Checking the Calibrant Level in your Agilent LC/MS System

    This Information Applies To: Agilent LC/MS System Issue Instructions on how to check the level of calibrant in your LC/MS system, and refill the calibrant bottle if necessary. Requirements Checking the level of calibrant should be performed weekly, or more often if frequent instrument tuning is required. Steps to Follow Open the door or remove the cover at the front of the instrument to reveal the calibrant bottle Carefully check that the intake tube that draws up calibration...
  • Filling the Calibrant Bottle in 6200/6400/6500 Series LC/MS Systems

    This Information Applies To: Agilent 6200/6400/6500 Series LC/MS Systems Issue Instructions on how to fill the calibrant bottle on 6200/6400/6500 Series LC/MS Instruments when calibrant levels are running low. Background The ability to autotune your Agilent LC/MS system relies on providing the instrument with the correct composition and concentration of calibration solution. Items Required The type of Calibrant used for autotune depends on the type of Ion Source installed on your...

Troubleshooting LC/TQ Contamination Issues

Answer

This Information Applies To the Following Products: All Agilent iFunnel and Agilent Jet Stream (AJS) 6400 Series LC/MS triple quadrupole systems, Agilent 6460A/C, 6470A, 6490A, and 6495A/B/C

 

Issue:

As LC Triple Quadrupole Mass Spectrometers (LC/TQ) have continued to become more sensitive to analytes, the capability to detect contaminants has also increased. The advent of Agilent Jet Stream ion source and iFunnel technologies starting with the Agilent 6490A, have contributed with the increasing sensitivity of the systems. It becomes more important to make sure that the entire flow path is clean. This article gives some specific examples of common contaminants found in LC/TQ systems and some tips on how to prevent and remove these types of contamination.

Note: It is uncommon to see all three contaminant classes described in the following sections, in the same instrument. However it is not uncommon to see one or two of these contaminant classes in combination in the same LC/TQ.

 

Steps to follow. 

To compare your LC/TQ instrument to the following examples, please implement MS2 scan data collection (also known as Solvent Blank Scan - no injection) to help identify background or contaminant issues. To ensure an accurate comparison, the chromatography (including gradient) and MS methods (including delta EMV) remain the same as the acquisition method you are currently using for your application.

  1. Create a solvent blank method by following these steps
    1. Save your current method with a different method name. Click Method > Save As, then input a suitable method name or add Solvent Blank Scan to the end of the method name.
    2. Set the acquisition mode in all segments to MS2Scan.
    3. Save the method.
    4. Run a solvent blank using this method.
        
  2. Compare Total Ion Chromatogram (TIC) in your solvent blank from step 1 to historical tune data of your instrument, including the checkout data produced at instrument installation.
    • Identify the main contaminant ions in the TIC data to try to identify potential common contaminants as per the examples in the following sections.

  3. Isolate potentially contaminated solvents by changing channel compositions as well as the solvents themselves. Isolate hardware (HPLC modules, LC analytical columns, etc.) by changing plumbing to bypass potential sources of contamination. Use a process of elimination to try to identify the source of these contaminants.

  4. Next, collect actual MS2 scan data using an LC gradient of 50:50 water(0.1% formic acid, FA): acetonitrile(0.1% FA) and compare main contaminant ions to the following examples. 

Inorganic Clusters - Example of Acetonitrile-Cu clusters

  • In this example, the TIC signal is above 5 x 108. As shown in Figure 1, m/z 145.0 has a height of 5673131. There are several classes of contaminants present; but, always address base peaks first. 

 

 Inorganic Clusters

Figure 1: MS2 Scan showing contamination in the low mass range. 

 

 

  • Inorganic acetonitrile-Cu clusters are present with 145 and 147 m/z as markers and causes ion suppression. When in doubt to the type of contaminant, apply collision energy (in either product ion scan or MRM modes for LC/TQ) to see product ions to determine structure.

Figure 2: Zoom of MS2 Scan showing a cluster of ions around m/z 145

 

 

 

Table 1: Mass of Inorganic Acetonitrile-Cu cluster ions
Mass of monoisotopic ion Ion Type Potential source of contaminant
144.98215 [(CH3CN)2+63Cu]+ Acetonitrile
146.98034 [(CH3CN)2+65Cu]+ Acetonitrile


Solution

  • Flush HPLC with EDTA solution (10 mM in water); Divert flow away from MS to waste. Use fresh Acetonitrile if necessary.

 

Polyethylene Glycol (PEGs) [C2H4O]nH2O·K

  • In this example, the TIC abundance is around 1.5 x 108 (see Figure 3). The prominent base peaks are Polyethylene Glycol or PEGs with markers as 173.1, 195.1, 217.1, 261.1, 305.1, etc. (+22 / +44 m/z apart). These ions cause high background and ion suppression at these levels. (See Figure 4)

 

Figure 3: Total Ion Chromatogram (TIC) showing abundance in the  1 - 1.7x 108 range.

 

Figure 4: MS2 Scan showing the major PEG contamination ions

 

 

Table 2: Mass of PEG contaminant ions
Mass of monoisotopic ion Ion Type Formula for subunit AnB shown in column 2
151.09649 [A3B+H]+ [C2H4O]nH2O
173.07843 [A3B+Na]+ [C2H4O]nH2O
195.12270 [A3B+Na]+ [C2H4O]nH2O
217.10465 [A4B+H]+ [C2H4O]nH2O
233.07858 [A4B+K]+ [C2H4O]nH2O

 

 

Solution

  • Remove analytical column.
  • Flush aqueous flow path with aqueous mobile phase containing no modifiers and at a high temperature (100C). This is only effective if the source of PEGs has been identified and removed from the flow path.

Possible Causes

  • Contaminants found in solvents, samples made/frozen in bad solvents, solvent bottles (if plastic), solvent bottle caps, and solvent bottle cap adhesive tape.
  • Identified in some packing material of LC analytical columns.  In this case, "wet" the column longer in 100% methanol.  

 

Triton and Triton Reduced (Detergent) [C15H30O][C2H4O]n

  • Now the background is almost clean after EDTA and aqueous flushing; but TIC is still around 1 x 108. 
  • Base peak is now 293.2 m/z and a marker for Triton or detergent. Interestingly enough, both inorganic and PEGs clusters also caused Triton ion suppression (See Figure 5).

  

Figure 5: MS2 Scan showing a reduction in contamination ions after EDTA and Aqueous flushing. Triton ions can still be seen.

 

 

Table 3: Triton and Triton Reduced (Detergent) ions
Mass of monoisotopic ion Ion Type Formula for M or subunit AB shown in column 2 Potential source of contaminant
293.24510 [AB1+Na]+ [C15H30O][C2H4O]n Triton, reduced
315.25299 [M+H]+ C18H34O4 Dibutyl sebacate, plasticizer

Solution

  • Wash bottles without detergent and dry in muffle furnace.
  • Can use acid etch and rinse thoroughly with aqueous followed by organic solvents (but not as effective as muffle furnace drying).

Possible Causes

  • Contaminants found in glassware washed with detergent.

 

 Bulb 
Learn how to effectively troubleshoot your Agilent 6400 Series LC/MSD system
LCMS-6400-1201s - Essential Housekeeping Tasks and Best Practices for Agilent 6400 Series LC/TQ e-learning course on Agilent University
 
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help