Hello Kirill_M ,
The averages for response or concentration are not available in the batch table. There have been enhancement requests made to add information like this to the batch table, but there is currently no timeline for when this will occur…
Support and educational resources for the PlateLoc Thermal Microplate sealer and PlateLoc Microplate Heat Seals .
Support Documents
PlateLoc Thermal Microplate Sealer Site Preparation Checklist
PlateLoc Thermal Microplate Sealer Installation Checklist…
For the GC-6890 FID detector issues you're experiencing, here's a simplified approach to troubleshoot the errors:
Gas Supplies:
Make sure all gas supplies (utility, makeup, hydrogen) are on.
Check that the pressure…
I see the clear difference between the split mode vs the splitless that now has the valid parameter. and the initial nitrogen elution seems a lot less problematic. thank you for your instructions!
Hi rguzmans1
To prepare a litre of this mobile phase, weigh 0.315g of ammonium formate and dissolve in 800mL of water, add 100µL formic acid and mix well, then make up to total volume of 1L.
Probably best to filter this after preparation through 0…
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue:
The Flame Ionization Detector (FID) jet could become clogged with ferrule material, column bleed, or heavy molecular analytes. The clogged jet will affect...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different high concentration, split analysis techniques including:
General use liners
Auto inject, general use
Auto...
This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet
Issue
Different liners are required for different splitless trace analysis techniques including:
General use liners
Pressure, temperature, volume (PTV)
Hot...
This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet
Issue
Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet...
This Information Applies To: Agilent GC Systems with Micro-ECD.
Issue
The Wipe Test is an annual test that performed by end user of the micro-ECD (Micro Electron Capture Detector or uECD) to help ensure radioactive safety when using the detector...
This Information Applies To: Agilent GC Systems with thermal conductivity detector (TCD)
Issue
Baking the thermal conductivity detector (TCD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items...
This Information Applies To: Agilent GC Systems with flame ionization detector (FID)
Issue
Baking out the flame ionization detector at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Items required...
This Information Applies To: Micro Electron Capture Detector (uECD)
Issue
Baking the micro electron capture detector (µECD) at high temperatures can remove sample contaminants that have deposited on the detector surfaces.
Warning ...
This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)
Issue
This article will describe how an operator can configure the makeup gas on both the instrument keypad and the software driver (Method Editor).
Background...
This Information Applies To: Agilent HPLC systems
Issue
The presence of ghost peaks can be a symptom of a system contamination issue, caused by accumulation of sample and/or foreign substances in the flow path. It is critical for high sensitivity...