Hello payninater ,
To utilize the inherited calibration curve you will need to copy the cal levels to the other targets. If you are doing this make certain your unknown targets have the same ISTD assigned. Quant does not verify this and simply applies…
Hello payninater ,
It is not possible to enter responses manually in the Method Editor.
As far as I know there should be no issue with setting your sample type and level in the batch table, creating a new method from acquired data, creating levels…
Hello HowngChan ,
For a dual needle multisampler the purge timer does not start until after the method with a different gradient or solvent system is loaded. By default, the worklist will have a 10 minute Wait Time for Ready. This can be set to a maximum…
Thanks for your reply. The syringe O rings are quite old so I will change them. I was able to overcome the issue by using a pulsed split mode on the inlet. Your other points are very helpful info and give me things to investigate.
Browse the help documentation by using the navigation menu or the links below.
Agilent Community Best Practices
Forum Best Practices
How to Post a Question
How to Reply to a Question
How to Attach Files
How to Configure Notifications…
Use the navigation menu to view our help articles.
Customize Your Email & Live Alert Notifications
Subscribe to Areas of the Agilent Community
Turn on or off Email Notifications
Which Electron Ionization Source Part 1 had questions. Which Electron Ionization Source Part 2 had some answers. Now, let’s talk about each source and tuning.
Stainless Steel/Inert vs. Extractor, Atune vs. Etune - The colored lines in the diagrams …
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power on the Agilent ICP-MS Foreline Pump when fully powering on the Agilent ICP-MS System, or after performing maintenance.
Background
The ICP...
This Information Applies To: Agilent ICP-MS MassHunter Software version 5.1 and above
Issue
The user manuals for Agilent ICP-MS Systems for MassHunter 5.1 and above are included in the ICP-MS MassHunter Help and Learning Center. There are manuals...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Elevated oxide ratio values are defined as greater than 1.5% for x-lens configuration or greater than 3% for s-lens configuration. Possible causes are contamination...
This Information Applies To: Agilent GC/MS Systems
Issue
The software version is a valuable information usually required for troubleshooting, software updates preparation or to check installed options.
Table of contents
MassHunter GC/MS Acquisition...
This Information Applies To: Agilent LC modules with RFID tag readers
Issue
Agilent uses RFID (radio frequency identification tag) tag readers in some HPLC components such as detector lamps and flow cells, pump heads, purge valve and mixers. However...
This Information Applies To: Agilent LC systems with Lab Advisor software.
Issue
In Lab Advisor, the Logs and Results register is available for gathering history and additional information on the system. It presents data collected from the configured...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
What information does the ICP-MS instrument status indicator provide?
Background
The ICP-MS Instrument Status Indicator on the instrument, and in the...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
How to power off Agilent ICP-MS systems before maintenance, for long-term shutdown, or for power cycling the instrument. Power cycling may be required when...
This Information Applies To: Agilent OpenLab CDS
Issue
Steps to close an instrument connection to a primary instrument controller, to start a successful connection to a secondary instrument controller, or to configure the instrument.
Background...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or replacement of consumable...