Hello,
Yes, that will not work. You need to test the length of the name as I said in the previous response. See the example below the unknowns have a low value of 1.56.
if(LEN(Compound_Name)>0, getconstant(“Relative response factor”),0.98)
…
UV-Vis spectroscopy is a well-established technique in many laboratories. For decades, laboratories have adapted to pain points associated with UV-Vis workflows such as daily instrument warm-up, yearly lamp replacement, and slow and cumbersome temperature…
Timmas You are running 7890A GC/5975 MS/ E.02.02 SP2 ? --- That version of MSD Chemstation was released November 2013 on Windows XP Professionals with SP3. The KBR error/fix you reference is from a document that does not support MSD Chemstation E.02.02…
Thanks Linh and Martin! first of all, it was not enabled in the instrument configuration. After the reconfiguration with intelligent reporting, I can see editing option! Also that course was very useful to me! Thanks a lot both!
I am having a similar issue. First, I replaced a needle seat and then the auto-referencing failed because I failed to recognize that the black sample holders needed to be removed. The needle mechanism seemed forever stuck in the parked position. I was…
Hello,
I am not sure what you did in your test, but it seems the alfresco instance is not running. The simplest issue might be the licensing if you renamed or changed the MAC on the VM guest.
Marty
We have an Agilent 1290 Infinity II with a multisampler and the instrument is connected to a mass spec. Last week while troubleshooting loss in signal, I noticed a leak in the autosampler. I determined that I needed to replace the needle seat. After replacing…
This Information Applies To: Agilent OpenLab CDS
Issue
Identifying the baseline codes and baseline code descriptions for OpenLab CDS
Background
OpenLab CDS Data Analysis includes the same integration algorithms as available in both OpenLAB...
This Information Applies To: Agilent OpenLab CDS
Issue
This article provides step-by-step instructions to filter a signal in an Intelligent Report.
Requirements
To be able to carry out the procedure as described, the privilege of Data Processing...
This Information Applies To: Agilent OpenLab CDS
Issue
This article outlines different ways to quantify the amount of compound in an unknown sample.
Background
Quantitation uses peak area or height to determine the amount of a compound in...
This Information Applies To: Agilent G1329A/B and G1313A Standard Autosampler
Issue
Replacing the needle seat of the Agilent 1100/1200/1260 Standard Autosampler may be required during regular maintenance, replacing a bent needle, or flow path blockage...
This Information Applies To: Agilent G1329A/B and G1313A Standard Autosampler
Issue
Replacing the needle of the Agilent 1100/1200/1260 Standard Autosampler may be required during regular maintenance, replacing a bent needle, or flow path blockage...
This Information Applies To: Agilent 1290 Infinity (G4226A) and Agilent 1260 Infinity (G1367E) HiP Autosamplers
Issue
Replacing the Needle Assembly of the Autosampler is required when:
The needle is visibly damaged.
Leaks or blockages in the...
This Information Applies To: Agilent 6000 Series LC/MS Systems equipped with MS40+ rotary vane pumps
Issue
Replacing the MS-40+ (rough pump) fluid periodically is a key maintenance procedure to improve vacuum performance and minimize the possibilities...
This Information Applies To: All Agilent 6000 Series LC/MS Systems without a gate valve for ion capillary installation/removal. For instructions for removing an ion capillary on LC/MS instruments with a gate valve installed, please click here
Issue...
This Information Applies To: Agilent 6000 series LC/MS instruments
Issue:
Cleaning the ion transfer capillary as part of regular maintenance or as part of troubleshooting when a loss of sensitivity in the instrument occurs.
Steps to follow ...
This Information Applies To: Agilent 6000 Series LC/MS systems.
Issue
Efficient electrospray ionization (ESI) in your LC/MS ion source requires a properly adjusted nebulizer needle in good condition. The nebulizer needle is a consumable item in...