Hello,
Here is an example report with a couple of options for the chromatogram object labeling. The first chromatogram is just the standard object. The second uses filtering to remove all the unknown peaks based on the compound name length. This removes…
Hello Tyler, Can you tell us some more? What are the concentrations you are working with? What settings are you using? It is highly likely that the recovery is low because of the sample preparation. Have you checked with an internal standard addition…
Hello,
We would need more information to diagnose the connection issues. If this is a workstation where both systems are connected directly, I would look at the power settings on the computer. Make sure the computer is not allowed to sleep or shutoff…
Brian,
The only thing you can do in Chemstation is see the signature of the ssizip file in ECM through the ECM properties panel. You cannot sign the files or reports from Chemstation, it has to be done from the ECM client. ECM signatures do not have…
Hi Jchavez
Have a look at this post ( https://community.agilent.com/technical/software/f/forum/6881/correction-factor-scaling-in-ol-from-ez-chrome/23614) that discusses how OpenLab and EZChrom data are scaled differently for different detectors. The…
The Digest of upcoming webinars has been updated as of January 23.
Notable additions include:
Food Testing (3)
Check Out:
Learn Why Atomic Absorption, MicroWave, and ICP Plasma Spectroscopy are the Perfect Tools for Both a Teaching and Research…
I finally figured out what was causing the communication error with MassHunter. It was one of the detector EPCs. Once I removed it from the instrument, MassHunter linked up without issue. Would have never imagined that an EPC would cause the issue, so…
Marty,
I need to apologize and thank you at the same time for your patience. I saw your screen grab in your previous response and realized what was the issue. I went back to your video and realized my mistake, too. It appears that the detectors are…
This Information Applies To: Agilent GC systems with Gas Clean Filter
Issue
If you have determined that the Gas Clean Filter is saturated, it mus be replaced (see the article When to Replace the Gas Clean Carrier Gas Filter ).
The steps described...
This Information Applies To: Agilent GC Systems 8890/8860/Intuvo 9000 /7890/7820/6850/6890
Issue
General guidance on GC column flow rates, outlining the correct column flow rate for optimal compound separation depends on the column internal diameter...
This Information Applies To: Agilent GC Method Translator
Issue
Detailed instructions on translating methods using the GC Method Translator.
Background
For ideal use of cases and applications with the GC Method Translator, see the article...
This Information Applies To: Agilent Solvent Vent Calculator
Issue
The Solvent Vent Calculator estimates the Agilent solvent vapor exit (SVE) accessory valve timing based on solvent type, oven temperature, vent flow, and pressure.
Background...
This Information Applies To: Use of the Agilent GC Method Translator
Issue
How to correctly use the GC Method Translator to convert a method to suit a different GC configuration or to optimize a method.
Background
The GC Method Translator...
This Information Applies To: GC Pressure Flow Calculator
Issue
Instructions for the correct use of the Pressure Flow Calculator.
Background
The Pressure Flow Calculator uses the column dimensions, oven temperature, and detector outlet pressure...
This Information Applies To: The Agilent GC Vapor Volume Calculator.
Issue
The Vapor Volume Calculator determines the vapor volume based on the injection volume, solvent type, inlet head pressure, and the inlet temperature. This calculated vapor...
This Information Applies To: Agilent GCs
Issue
Agilent provides free GC calculators and tools to assist with method development.
Background
Tips : The calculators are available for downloading at Agilent.com .
Vapor Volume...
This Information Applies To: Agilent GC Systems
Issue
The Agilent Gas Clean Carrier Gas Filter delivers faster stabilization times for increased GC productivity, removes oxygen, moisture, and hydrocarbons from the carrier gas for MS applications...
This Information Applies To: Agilent GC systems
Issue
When a GC heater cannot reach setpoint, or taking too long to reach setpoint, the GC will enter a "Thermal Shutdown" (or similar heater error) state to protect itself. This state will turn off...