Thanks Linh and Martin! first of all, it was not enabled in the instrument configuration. After the reconfiguration with intelligent reporting, I can see editing option! Also that course was very useful to me! Thanks a lot both!
I am having a similar issue. First, I replaced a needle seat and then the auto-referencing failed because I failed to recognize that the black sample holders needed to be removed. The needle mechanism seemed forever stuck in the parked position. I was…
Hello,
I am not sure what you did in your test, but it seems the alfresco instance is not running. The simplest issue might be the licensing if you renamed or changed the MAC on the VM guest.
Marty
We have an Agilent 1290 Infinity II with a multisampler and the instrument is connected to a mass spec. Last week while troubleshooting loss in signal, I noticed a leak in the autosampler. I determined that I needed to replace the needle seat. After replacing…
The Digest of upcoming webinars has been updated as of June 17.
Did you miss the GC Live Event? Don't worry, check out the recording here: GC Live is Back in June with the Next Big Little Thing in GC: The New 8850 GC System !
Notable additions include…
Best wishes for success! Applications training is dependent upon the services available in your country. I suggest you contact your local Agilent sales person or dealer for more information specific to your country.
This Information Applies To: Agilent Liquid Chromatography (LC) systems
Issue:
Shutting down the HPLC for a long period requires following this general procedure to prevent damage to the system and the column.
Steps to follow:
Store the...
This Information Applies To: Agilent G7129A/B Autosampler with Sample Thermostat/Cooler G7167-60101 or G7167-60005.
Issue:
The autosampler may stop working from time to time, potentially giving off loud noises and switching off the light. The only...
This Information Applies To: Agilent 5977A/B GC/MSD with EI source, EI Extractor Source, or CI Source.
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte...
This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD
Issue:
During normal operation of the GC/MSD, the source will require cleaning. It is time to clean the source when:
There is a loss in analyte response not improving...
This Information Applies To: Agilent 5977 and 5975 Series GC/MSD systems
Issue
The autotune fails and reports the error Tune action stopped: Gain and offset = 0 and peaks are too narrow!
Resolution
The MSD autotune program is run after routine...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Condition new or cleaned ICP-MS skimmer cones and sampler cones by aspirating a Conditioning Solution before use.
Background
Drift up can be a sign of...
Turn On the GC/MS Interface and Oven Using OpenLab CDS
Multiple Results
This Information Applies To: Agilent GC/MS instruments controlled by OpenLab CDS software
Issue
Turning on the GC/MS interface heater and Oven using OpenLab CDS software.
Steps to Follow
From the Home tab, click Method > Instrument Setup...
This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900)
Issue
The signal monitor function in MassHunter can help with troubleshooting low sensitivity, carry over, contamination, or high background. It can also be used to optimize...
This Information Applies To: Agilent UV detectors (Variable Wavelength Detectors, Multiple Wavelength Detectors, and Diode Array Detectors)
Issue
Checking accumulated hours on the UV lamp.
Steps to Follow
There are four ways to check the accumulated...
This Information Applies To: Agilent 7700/7800/7850/7900/8800/8900 ICP-MS
Issue:
Poor internal standard (ISTD) mixing can cause high relative standard deviation (RSD).
The recommendations in this article are applicable if:
It takes a long...