We've been having issues with our LC-MS/MS 1260/6490 and 1290/6470 systems giving "LC Offline" errors when the instrument is all ready to go and we start a worklist. It clearly is connected, because it's showing the LC status the entire time. We're using…
From a GCMS perspective, 1.9 ml/min into the ion source is not optimal. The MS prefers down around 1.2 ml/min. Sensitivity is better and spectral tilt is more typical.
Septum head or septumless head? What is your gas supply pressure? When was the last…
Hello,
Question 1: Are you opening a snapshot or using review completed injections to open the in-progress injections? Is it possible you were on a replicate injection when trying to change the type.
Question 2: Andy is correct you must have the…
Helllo mighty Marty,
we have only one Infinity II device, so it can be useful. It works same as your example. But if we are going to compare this system with GC catalog it seems to be very painful to write all the info again and again when you are going…
Hello payninater ,
To utilize the inherited calibration curve you will need to copy the cal levels to the other targets. If you are doing this make certain your unknown targets have the same ISTD assigned. Quant does not verify this and simply applies…
Hello payninater ,
It is not possible to enter responses manually in the Method Editor.
As far as I know there should be no issue with setting your sample type and level in the batch table, creating a new method from acquired data, creating levels…
Hello HowngChan ,
For a dual needle multisampler the purge timer does not start until after the method with a different gradient or solvent system is loaded. By default, the worklist will have a 10 minute Wait Time for Ready. This can be set to a maximum…
This Information Applies To: Agilent 8890/8860 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...
This Information Applies To: Agilent 6000 Series LC/MS systems equipped with MS40+ rotary vane pumps
Issue
Instructions on how to replace the filter in an Agilent MS 40+ pump as part of annual pump maintenance
Background
The filter in the...
This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900) with ICP Expert 7.5.x Software
Issue
The Recommended Routine Maintenance Schedule does not adjust for the number of samples analyses or the sample matrix, both of which...
This Information Applies To: Agilent 7890/7820 GC Systems with SSL, MMI, PTV, or VI inlets.
Issue
Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure...
This Information Applies To: Agilent LC pumps
Issue:
Stuck or obstructed check valves (Active Inlet Valve (AIV), Passive Inlet Valve (PIV), Outlet Ball Valve (OBV)) can cause issues such as no-flow or no-pressure, moderate to strong drops in pressure...
This Information Applies To: Agilent 1260 Infinity II multisampler, 1290 Infinity II multisampler, 1260 Infinity II SFC multisampler, and 1260 Infinity II bio-Inert multisampler
Issue
The needle bends while moving to the wash port or showing one...
This Information Applies To: Agilent OpenLab CDS 2.x
Issue
This article outlines the step-by-step instruction on how to overlay chromatograms for OpenLab CDS 2.x
Steps to Follow
Load data from result sets or single samples. See Load or add...
This Information Applies To: Agilent Lab Advisor Software, Agilent Multicolumn Thermostat
Issue:
Agilent Lab Advisor is a great tool to diagnose the issue of your HPLC. Running the tests that are relevant to your specific issue in the Service &...
This Information Applies To: Agilent OpenLab CDS (ver 2.x)
Content:
The following procedure shows how to review completed injections in a running sequence in OpenLab CDS (ver 2.x).
Steps to Follow:
Select the sequence from the Run Queue...