Update: FID output signal was very low even when trying to ignite, so we checked the interconnect spring and noticed it was shorter than the ones in the other GCs (not that close to the detector base; dont know if it was always like that or something…
Hi,
We are struggling to solve an issue with an Agilent 8860 GC-FID we purchased in 2019. It is used to quantify propane in headspace samples that are manually injected. It is possible but unlikely that some of those headspace samplings result in the…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Tip for other users that may encounter an FID Ignition problem: try blowing on the FID when it is trying to light. If the FID lights - there is an issue with the gas flows. It is best to use a flow meter like the ADM Flow Meter for gas chromatography…
Hey. Johnson we solved the Problem by removing the igniter Castor with collector nut and the spring locks Info the right place. Another Problem could be the malfunction electrometer.
I have a 5890 GC running ASTM D2887, simulated distillation. It is equipped with a FID and recently we noticed the FID won't light. Flows are good and gases are good. Looking at the setpoint/actual reading, it states it is at the correct temperature …
I completed the inlet leak test and it passed and I also completed the FID jet restriction test and it passed. Yet, when I compare scans on the 8890 the peaks shift more compared to the same injections on my 6890. Does anyone have any thought on this…
Hello DET ,
Sorry for the late response on this.
What sort of spiking do you see usually? What is the output of the FID, in pA?
Are you using any UPS systems, or are the GCs plugged directly into the wall?
This Information Applies To: Agilent 4200 TapeStation 4150 TapeStation, TapeStation controller software
Issue
The system verification can be run to check the TapeStation performance, for example, when you suspect an instrument hardware fault.
...
This Information Applies To: Agilent 4200 and 4150 TapeStation software.
Issue
How to uninstall and reinstall the 4200 and 4150 TapeStation software.
Background
An uninstallation and reinstallation of the TapeStation software should be performed...
This Information Applies To: 2200 TapeStation instruments, 2200 TapeStation controller software
Issue:
The Agilent 2200 TapeStation controller software won't launch but displays the correct serial number in the top-right corner of the loading screen...
This Information Applies To: 4200, 4150 and 2200 TapeStation, TapeStation analysis software A02.02 or higher.
Issue:
When using the Agilent 4200, 4150 and 2200 TapeStation systems, the DV 200 of FFPE RNA samples can be calculated within the TapeStation...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, ProSize Data Analysis Software
Issue
How to export a ladder as a size calibration file and import it into a different run in ProSize.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
It is possible to set up the 2100 Expert software so that the files are automatically exported each time a chip run is complete.
Steps to Follow
To turn...
This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software
Issue:
How to identify and troubleshoot late migration of the ladder, marker, and sample peaks in data generated on the Agilent 2100 Bioanalyzer system.
Background
...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
If communication between the 2100 Bioanalyzer and the PC cannot be established, or if an intermittent communication loss occurs, follow the instructions in...
This Information Applies To: Agilent 2100 Bioanalyzer system, 2100 Expert software
Issue
Instructions on how to import from or export data into LIMS software.
Steps to Follow
Data import
When importing sample names and sample comments from...