This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900)
Issue
Description of ICP-OES 5000 Series Status Indicators (LEDs)
The Agilent 5000 Series ICP-OES (5100/5110/5800/5900) has two different LED indicators to advise the status of the instrument. The state of both these indicators determines if the instrument is in a fault condition or ready to operate condition. The ready to operate condition is shown when both indicators are a steady green color. Any other color (steady or flashing) indicates a fault condition.
On/Off Switch green LED |
Instrument State |
Brief flash once every 10 seconds |
Most of the instrument is unpowered. The instrument can be turned on (powered up) by depressing the soft on/off switch at the front. Mains power is active inside the instrument |
Flashing slow (1 Hz) |
Mains power module is proceeding through the startup check sequence while transitioning between Off and On states |
On |
Instrument in On |
Flashing fast (2 Hz) |
Mains power module failed its power-up self-check, instrument is unusable. Mains power supply is not adequate or the internal mains module requires replacement |
2. Indicator two. Instrument status LED indicator, on the top-right side of the instrument. The LED color combination provides feedback on instrument status as indicated in Table 2.
Status LED Color |
Instrument State |
None (off)
|
• Instrument off • Instrument starting • Instrument switching off/shutdown Note 1: Soft on/off switch LED indicates status of any powering up activity during these states |
Light Purple |
• Instrument Startup |
Red Slow Pulse .. .. .. |
• Firmware Loading |
Yellow flashing (0.5 Hz) .. .. .. |
• Host is not connected (regardless of whether plasma is lit or not) Note 2: This LED status has higher priority than solid Yellow or Green, but lower than Red |
Yellow
|
• Host is connected but the instrument is not ready to begin a plasma ignition sequence. Possible causes include door or torch handle interlock tripped, low instrument water flow (uses camera flow sensor to detect if water connected and flowing) |
Green |
• Host is connected, ready to attempt a plasma ignition sequence • Host is connected, plasma ignition sequence is underway • Host connected and plasma lit and running. Possible to run analysis of samples |
Red |
• Instrument (FW) error causing a halt to analysis or restricting normal instrument operation, user intervention required to rectify the problem and regain error free status • Note 4: this LED status has the highest priority (supersedes all others) |
Learn how to effectively maintain and operate your Agilent ICP-OES System: