This Information Applies To: Agilent Single Quadrupole GC/MS Instruments
Issue
During data acquisition, users may override the Solvent Delay state for the current run.
Steps to Follow
Figure 2. Run Queue Display
1. Run in the Acquiring State
Caution: The selected run must be in the Acquiring state.
|
3. Click Instrument Status in the top ribbon to open the Instrument Status screen, then click the Dashboard tab. Figure 3.
Figure 3. Dashboard tab in Instrument Status window
1. Instrument Status button, 2. Dashboard tab under Instrument Status screen
Tips: This option will be available after the sample injection. |
Figure 4.Option under MSD window
1. MSD Window under Dashboard, 2. Option to Override Solvent Delay
Learn how to effectively operate your Agilent GC/MSD:
Agilent 5977 GC/MSD with OpenLab CDS eLearning course available from Agilent education |