LC Articles Creating a System Report PDF with Agilent Lab Advisor
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • -LC Operation
      • How to Install Lab Advisor
      • How to Add a New System in Lab Advisor
      • Creating a System Report PDF with Agilent Lab Advisor
      • How to Check, Configure and Use the IP Address of an Agilent HPLC Module
      • LED Status Indicators
      • Installing the HPLC Advisor App
      • Using the Method Translator Calculator in the HPLC Advisor App
    • +LC Maintenance
    • +LC Troubleshooting
    • +LC Best Practices
    • +LC Firmware

Creating a System Report PDF with Agilent Lab Advisor

Created by Max Schütte Max Schütte over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 10 months ago

This Information Applies To: Agilent Lab Advisor Software for any Agilent LC System.


Issue:

To successfully troubleshoot an Agilent LC system or module, the System Report is an essential resource. Agilent Support representatives may ask you to provide them with this information.

The report contains detailed summary information of your system configuration, modules, error messages and logs, firmware versions, LAN settings, and past test results. This article shows you how to create it and what to be aware of.


Steps to follow:

  1. Go to your Lab Advisor Software and connect to your instrument (see How to add a new system in Lab Advisor and How to Install Lab Advisor).
  2. On the left side navigation bar, under My Instrument (as per Figure 1; this is the name of your instrument), click System Report.
  3. Enter your contact information (optional) and, if necessary, comment on any issue or for later reference.
  4. Make sure that the check boxes for Include PC information in the General tab and for Include instrument actuals in the My Instrument tab are selected.
  5. Important: In the General tab, under Included Information, typically select Include last 10 tests of module and last 50 log entries if not asked differently. If you want to be on the safe side, select Include all data.
  6. In the bottom right, click Create Report.
  7. On the preview that appears in the top menu, click the Save Icon   to export the Report as a PDF file, adjusting the documents name if necessary.
  8. If asked, send the file to your local Agilent Support representative.

The following animation will guide you through the instructions on how to generate the report and include all necessary information (Figure 1):


Figure 1. Generating the Lab Advisor System Report and saving it as a PDF file

 Tip 
Learn more on how to effectively operate the Agilent Lab Advisor Software:
Agilent Lab Advisor User Manual on Agilent.com
  • Troubleshooting
  • Instrument Control Software
  • Lab Advisor
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help