Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles Troubleshooting MCGV Cross Port Leaks in Quaternary Pumps
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • -LC Pumps Troubleshooting
        • Troubleshooting unstable retention time issues – Cleaning the multichannel gradient valve (MCGV) on Agilent quaternary pump modules
        • Troubleshooting MCGV Cross Port Leaks in Quaternary Pumps
        • How to Troubleshoot no flow or no pressure issues: Binary/Quaternary/Isocratic Pumps - G131XX, and G711XX
        • No Pressure, No Flow, RT shift Issues: Cleaning Acetonitrile (ACN) Aggregations in Agilent Pumps
        • Troubleshooting Agilent Pump Errors due to Blockage of a Pump Head
        • How to Release a Stuck Inlet Valve in a G7120A High Speed Pump
        • Troubleshooting "Vacuum Limit Not Reached" Error in Agilent LC Pumps
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • +LC Detectors Troubleshooting
    • +LC Best Practices
    • +LC Firmware

Troubleshooting MCGV Cross Port Leaks in Quaternary Pumps

Created by Shanice Rakotonindraina Shanice Rakotonindraina 2 months ago | Last modified by Carlos Vargas Carlos Vargas 1 month ago

This Information Applies To: Agilent 1100, 1200, 1290, and 1260 Quaternary Pumps


Issue

The Multi-Channel gradient valve may create pressure fluctuations and retention time shifts due to cross port channel leaks. To check if there are cross port leaks in the MCGV, the "bubble test” is an easy and quick way to confirm it. 

The cross port leak will result in solvents from unintended channels to mix with the mobile phase and will result in retention time shifts as well as pressure fluctuations.


Steps to Follow

  1. While pumping from each channel, raise the solvent filters above the liquid to introduce an air bubble in each line. Allow it to go past the bottle cap and on the downward slope of the solvent line.
  2. Place the bottle filters back into the solvent in each bottle. It’s a good idea to mark the bubble in each line with a sharpie or tape. If using a piece of tape, suggest marking the top of the air bubble to make it easier to see if it moves.
  3. While pumping at 1 ml/min from channel A, look for movement of bubbles in the other three lines. If the bubble moves past the mark or tape in the other three lines, it means you have cross port leaking of that valve in the MCGV.
  4. Repeat the process for 100% B, 100% C, and 100% D. Look for bubble movement each time to indicate MCGV cross port leaks.

If the cross port leak is determined, contact your local Agilent Support Representative for further support.

 Tip 
Learn how to effectively use your Agilent Multi-Channel Gradient Valve:
Good practices for use of the Multi-Channel Gradient Valve in Agilent quaternary pump modules  on Agilent Community
  • Troubleshooting
  • 1100 series
  • 1290 infinity
  • Pumps
  • 1260 Infinity
  • 1260 Infinity II
  • 1200 Series
  • Analytical LC
  • quaternary pump
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help