Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles What Components determine Gradient Composition and Flow Accuracy in Binary and Quaternary Pumps
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  24 Mar 2023
  • Modified :  27 Jul 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Pumps
  • Product Name :  1100 Series LC System 1120 Compact LC 1200 Series LC System 1220 infinity II LC System 1260 Infinity II Bio-Inert LC System 1260 Infinity II LC System 1260 Infinity II Prime Bio LC System 1260 Infinity LC System 1260 Infinity II Prime LC System 1290 Infinity II Bio LC System 1290 Infinity II LC System 1290 Infinity LC System
  • Task :  Best Practices Troubleshooting
Related Articles
  • How to Troubleshoot no flow or no pressure issues: Binary/Quaternary/Isocratic Pumps - G131XX, and G711XX

    This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps Issue: How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps. Resolution: The following information should help you find what part of the pump is causing a flow or pressure problem. Figure 1 and Figure 2 show the hydraulic paths of the pumps. Figure 1 - Hydraulic path for the Isocratic and quaternary pump...
  • Reinstalling a pump head on Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules

    This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Modules Issue: Reinstallation of the pump head on the pump metering drive during maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module Steps to follow 1. Ensure that your Agilent 1260 Infinity Pump module is turned off at the front power switch. 2. Place your pump head on the pump metering drive. 3. Insert the two pump head bolts through the two holes in the pump head...
  • How to Remove the Pump Head From an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump

    This Information Applies To: Agilent 1260 Infinity Binary, Quaternary or Isocratic Pumps. Issue: Removal of the pump head from the pump metering drive for maintenance on an Agilent 1260 Infinity Binary, Quaternary or Isocratic Pump Module. Steps to follow Caution : When loosening the pump head blots, make sure that the bolts are loosening evenly left and right. Do not loosen one pump bolt fully while leaving the other bolt tight as it may cause misalignment of the pump head...
  • General Recommendations for Using an Agilent RID

    This Information Applies To: Agilent Refractive Index Detectors, G1362A and G7162A/B. Issue: The RID is sensitive to its environment and can experience stability problems when exposed to contamination, vibrations, mechanical shocks, or changes in temperature or pressure. Resolution: Considerations on the general set-up: To avoid cross-talk between different solvent channels, preferably use an isocratic pump when operating a RID. If it is not possible to use an isocratic pump, consider...
  • Modifying Solvent Compressibility for LC Pumps in OpenLab CDS

    This Information Applies To: Agilent OpenLab CDS, G7110B Isocratic Pump, G7111A/B, G5654A Quaternary Pumps, and G7112B Binary Pump. Issue Incorrectly applied solvent compressibilities can result in an increased pressure ripple, or inaccurate solvent composition in LC pumps. Background In HPLC, liquids under high pressures are pumped through the system, which can result in measurable liquid compression. Though the amount of compression is very small, it is enough to change the flow rate...

What Components determine Gradient Composition and Flow Accuracy in Binary and Quaternary Pumps

Answer

This Information Applies To: Agilent Binary and Quaternary LC Pumps


Issue

When performing a gradient analysis, it’s important to know what component determines composition if it becomes necessary to troubleshoot chromatographic inconsistencies. Those inconsistencies can manifest themselves as peak retention time shifts (late or early peaks) and merged or missing peaks. Troubleshooting strategy depends on the pump type being used. This article will describe how binary and quaternary pumps create gradient runs.


Resolution

  • Binary Pump
  • Quaternary Pump

Binary Pump

Figure 1 shows the front of a typical binary pump.


Binary pump
Figure 1. Binary Pump
1. Degasser 2. Solvent Selection Valve, 3. Pressure Sensor, 4. Mixer, 5. Damper, 6. Purge Valve, 7. Pump B, 8. Pump A, 9. Inlet Check Valve, 10. Outlet Check Valve

 

Figure 2 shows the flow schematic within the binary pump.


Binary pump schematic
Figure 2. Binary Pump Schematic
1. Seal Wash Solvent, 2.Seal Wash Pump, 3. Pump A, 4. Solvent Selection Valve, 5. Solvents, 6. Purge Valve, 7. Pump B, 8. To Sampler, 9. Pressure Sensor/Damper, 10. Inlet Check Valve, 11. Outlet Check Valve, 12. Mixing Chamber, 13. Mixer

Binary pumps have two pump heads and are referred to as Pump A or Channel A and Pump B or Channel B. If there is an SSV (solvent selection valve), then the user can choose either A1 or A2 for Pump A and B1 or B2 for Pump B. The user needs to select either A1 or A2 to be mixed with either B1 or B2. If there is no SSV, then there will be only one solvent line for Pump A and one for Pump B. Some pumps may have a built-in degasser, as this example shows, while others may have an external degasser.

By convention, Pump A will flow with aqueous solvent while Pump B will be for organic solvent. So, if a user sets a flow at 1 mL/min and wishes to have a 50:50 composition for the mobile phase, both Pump A and Pump B will be delivering solvent at 0.5 mL/min. If the user wants 100% organic, then Pump B will be flowing at 1.0 mL/min while Pump A is idle and vice versa. In this way, the user can program any composition change during the gradient run with a timetable.

Notice that Pumps A and B determine both the Flow Rate as well as the Composition Accuracy.

The solvents in a binary pump are mixed after the pump heads and is sometimes referred to as high pressure mixing. An advantage is that if there is outgassing, such as mixing water and methanol, it will be suppressed since it is under system pressure.

The simplest way to troubleshoot in case of retention time shifts is to flow 100% off Pump A and measure the flow accuracy in a volumetric flask. Repeat with Pump B. If the Flow Rates are accurate, then the Composition Accuracy will be accurate as well. The SSV doesn’t determine the flow accuracy so there is no need to test both channels off each pump.

 

Quaternary Pump

Figure 3 shows the front of a typical quaternary pump.


Quaternary pump
Figure 3. Quaternary Pump
1. Degasser, 2. Multichannel Gradient Valve, 3. Pump Head, 4. Inlet Check Valve, 5. Purge Valve, 6. Outlet Check Valve

Figure 4 shows the flow schematic within the pump.


Quaternary pump schematic
Figure 4: Quaternary Pump
1. Solvents 2. Degasser, 3. Seal Wash Solvent, 4. Seal Wash Pump, 5. Multichannel Gradient Valve, 6. Pump Head, 7. Purge Valve, 8. Pressure Sensor/Damper, 9. Outlet Check Valve, 10. Inlet Check Valve

Quaternary pumps have a single pump head. They also have a device called Multichannel Gradient Valve (MCGV) where the user can select up to four different solvents. It is the device that will allow the user to select two or more channels to mix using solenoid valves.

So, if a user sets the flow at 1 mL/min and wishes to have a 50:50 composition from channels A and B, the pump head will flow at 1 mL/min while the MCGV will cyclically alternate between channels A solenoid valve 50% of the time and channel B solenoid valve 50% of the time. If the user wants 100% from channel A, then solenoid valve A will be fully open while the others are closed.

Notice then, that for a quaternary pump, the Composition Accuracy is determined by the MCGV while the Flow Rate is determined by the pump head.

When the solenoid valves open, slugs of solvent will move along to the pump head. In theory, there isn't a barrier between the solvent slugs as they move to the pump head, so one can say that mixing begins when the solvents leave the MCGV. For this reason, quaternary pumps are referred to as low pressure mixing as the MCGV operates at atmospheric pressure. True mixing, however, happens in the pump head where the movement of the pump piston will create turbulence and thoroughly mixes the solvents.

To troubleshoot the pump, set a flow through use a single channel and measure the flow accuracy using a volumetric flask. If the Flow Rate is accurate, then the issue may be due to the MCGV not mixing properly. To determine if that is true, then the MCGV Compositional Accuracy needs to be tested. See the article Quick Way to Manually Determine Gradient Composition Accuracy.

Therefore, for Binary Pumps, if the Flow Rate is accurate, then the Composition Accuracy will also be accurate. For Quaternary Pumps, both the Flow Rate and Composition Accuracy must be accurate.

 