Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles Troubleshooting MCGV Cross Port Leaks in Quaternary Pumps
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  6 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems
  • Component :  LC Pumps
  • Product Name :  1100 Series LC System 1200 Series LC System 1260 Infinity II LC System 1260 Infinity LC System 1260 Infinity II Prime LC System 1290 Infinity II LC System
  • Task :  Troubleshooting
Related Articles
  • Troubleshooting Unstable Retention Time Issues - Cleaning the Multichannel Gradient Valve (MCGV) on Agilent Quaternary Pump Modules

    This Information Applies To: All Agilent LC pump modules with MCGV installed (quaternary pumps) Issue: Shifting retention times: The composition of the mobile phase in liquid chromatography directly influences retention times. If any of the channels of the MCGV are partially or fully blocked, the gradient produced will be different from the analytical method. Clogging of the MCGV may occur when the system is not rinsed properly or stored incorrectly. If the following steps don’t solve the...
  • Good Practices for Use of the Multi-Channel Gradient Valve in Agilent Quaternary Pump Modules

    This Information Applies To: Agilent Quaternary Pumps with Multichannel Gradient Valves Issue: The correct use and maintenance of multichannel gradient valves (MCGVs) in Agilent Quaternary Pump modules will prevent blockage or damage of this assembly. It will help ensure the MCGV functions correctly and reliably. Resolution : Warning : Use the proper PPE and precautions when handling hot water and solvents. General: Before operating the quaternary pump, flush the...
  • How to Troubleshoot Low Pressure or No Flow/Solvent Coming Out of 1100/1200 Series Isocratic/Quaternary/Binary Pumps

    This Information Applies To: Agilent 1100/1200 Series G1310A Isocratic Pump, G1311A Quaternary Pump, and G1312A/B Binary Pump modules Issue This article shows different options to troubleshoot low pressure or no flow issues in 1100/1200 series Isocratic/Quaternary/Binary pumps. Items Required Open-end wrench 14 mm (p/n 8710–1924 ) Open-end wrench 1/4 - 5/16 inch (p/n 8710–0510 ) Pair of Tweezers Steps to follow The following steps will guide you to find out which part is causing...
  • What Components determine Gradient Composition and Flow Accuracy in Binary and Quaternary Pumps

    This Information Applies To: Agilent Binary and Quaternary LC Pumps Issue When performing a gradient analysis, it’s important to know what component determines composition if it becomes necessary to troubleshoot chromatographic inconsistencies. Those inconsistencies can manifest themselves as peak retention time shifts (late or early peaks) and merged or missing peaks. Troubleshooting strategy depends on the pump type being used. This article will describe how binary and quaternary pumps create...
  • How to Troubleshoot no flow or no pressure issues: Binary/Quaternary/Isocratic Pumps - G131XX, and G711XX

    This Information Applies To: Agilent G1311A/B/C, G7111A/B, Quaternary Pumps, and G1310A/B, G7110B Isocratic Pumps, G1312A/B/C, G7112B Binary Pumps Issue: How to diagnose no flow or no pressure issues on binary, quaternary, or isocratic pumps. Resolution: The following information should help you find what part of the pump is causing a flow or pressure problem. Figure 1 and Figure 2 show the hydraulic paths of the pumps. Figure 1 - Hydraulic path for the Isocratic and quaternary pump...

Troubleshooting MCGV Cross Port Leaks in Quaternary Pumps

Answer

This Information Applies To: Agilent 1100, 1200, 1290, and 1260 Quaternary Pumps


Issue

The Multi-Channel gradient valve may create pressure fluctuations and retention time shifts due to cross port channel leaks. To check if there are cross port leaks in the MCGV, the "bubble test” is an easy and quick way to confirm it. 

The cross port leak will result in solvents from unintended channels to mix with the mobile phase and will result in retention time shifts as well as pressure fluctuations.


Steps to Follow

  1. While pumping from each channel, raise the solvent filters above the liquid to introduce an air bubble in each line. Allow it to go past the bottle cap and on the downward slope of the solvent line.
  2. Place the bottle filters back into the solvent in each bottle. It’s a good idea to mark the bubble in each line with a sharpie or tape. If using a piece of tape, suggest marking the top of the air bubble to make it easier to see if it moves.
  3. While pumping at 1 ml/min from channel A, look for movement of bubbles in the other three lines. If the bubble moves past the mark or tape in the other three lines, it means you have cross port leaking of that valve in the MCGV.
  4. Repeat the process for 100% B, 100% C, and 100% D. Look for bubble movement each time to indicate MCGV cross port leaks.

If the cross port leak is determined, contact your local Agilent Support Representative for further support.

 Tip 
Learn how to effectively use your Agilent Multi-Channel Gradient Valve:
Good practices for use of the Multi-Channel Gradient Valve in Agilent quaternary pump modules  on Agilent Community
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!