Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Liquid Chromatography Portal
  • Agilent Knowledge Portal
Liquid Chromatography Portal
LC Articles Troubleshooting "Vacuum Limit Not Reached" Error in Agilent LC Pumps
  • LC Articles
  • LC Videos
  • LC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Liquid Chromatography
    • +LC Operation
    • +LC Maintenance
    • -LC Troubleshooting
      • Troubleshooting High Pressure Issues in Agilent LC Systems
      • Troubleshooting Peak Area Variation Between Injections
      • Forced Cold Start Procedure
      • Troubleshooting Remote/Cluster Partner Missing Error
      • How to Access Diagnostic Buffer File from LC Modules for Advanced Troubleshooting
      • How to Free up Licenses After License Check Message on Agilent Lab Advisor Software
      • "Could not open channel RD" error message in Lab Advisor
      • -LC Pumps Troubleshooting
        • Troubleshooting unstable retention time issues – Cleaning the multichannel gradient valve (MCGV) on Agilent quaternary pump modules
        • Troubleshooting MCGV Cross Port Leaks in Quaternary Pumps
        • How to Troubleshoot no flow or no pressure issues: Binary/Quaternary/Isocratic Pumps - G131XX, and G711XX
        • No Pressure, No Flow, RT shift Issues: Cleaning Acetonitrile (ACN) Aggregations in Agilent Pumps
        • Troubleshooting Agilent Pump Errors due to Blockage of a Pump Head
        • How to Release a Stuck Inlet Valve in a G7120A High Speed Pump
        • Troubleshooting "Vacuum Limit Not Reached" Error in Agilent LC Pumps
      • +LC Injectors Troubleshooting
      • +LC Column Compartments Troubleshooting
      • +LC Detectors Troubleshooting
    • +LC Best Practices
    • +LC Firmware

Troubleshooting "Vacuum Limit Not Reached" Error in Agilent LC Pumps

Created by Max Schütte Max Schütte 6 months ago | Last modified by Carlos Vargas Carlos Vargas 1 month ago

This Information Applies To: Agilent G1311B/C, G7110B, G7111A/B, G7112B, and G7120A Pumps and Agilent G7122A, G4225A, G1379A/B, and G1322A Degassers


Issue

The error message "Vacuum Limit Not Reached" will appear if the degasser does not reach a preset limit (default value is 100 mbar) in a predefined time window (default 8 minutes after powering on the module).  This may happen for many reasons.

One reason may be due to the condensation of vapors from volatile organic solvents. The vapors can condense in any part of the degasser and not restricted to any component. As such, the vacuum chamber, vacuum tube or the vacuum pump may all experience vapor buildup. If this occurs in the vacuum pump, it makes it less efficient to remove the air and achieve the 100 mbar level in time. It doesn’t happen often but the vacuum pump can totally fail and not turn on at all. This is easy to determine as no sound will be heard when the module is powered on.

The vapor can also cause cracks or fissures to develop in the vacuum tubing thus making it impossible to ever achieve the preset vacuum level.

The vacuum chamber may also develop condensate buildup, which will require a longer time to evacuate thus giving the error. In some rare cases, the chamber may even develop a crack.

In addition, other internal components can fail in rare cases, but in this case it is better to contact an Agilent Support representative.

The external degassers will not give "Vacuum Limit Not Reached" messages as they aren’t connected to Controlling software or Lab Advisor. The red error light will go on to indicate a problem. In addition, if a remote cable is connected to the pump, it will also turn all the modules red. Whether this cable is connected or not is optional.


Steps to follow

  • To avoid this problem when using a built-in degasser, we recommend not to turn off the pump at all.
  • If you must power off the pump, make sure to flush out the volatile solvent with isopropanol (IPA) thoroughly before doing so. In case, there is condensate buildup, which causes a vacuum timeout, an evacuation mode is available for 1260 Infinity II and 1290 Infinity pumps to recover from this problem (not applicable for external degassers and 1260 Infinity pumps). Just use the latest Lab Advisor, connect to the instrument, navigate to Instrument Control, toggle the evacuation mode on as shown in Figure 1, and follow the onscreen instructions:




Figure 1: Where to start the Evacuation Mode with Lab Advisor.

  • For internal and external degassers alike, another option is to switch on the module, wait until the degasser goes into error and then restart the module. Doing so a couple of times may be sufficient to remove any condensate that may have formed. 
    Note: The G1322A has some advantages over internal degassers for solvents with low boiling point. The tubing is not that porous and the vacuum level is not as high as the internal degassers, thus less solvent vapor will condensate in the chamber. The G1322A also has two different modes of action, even if there is condensate, it will be removed more efficiently.

If you cannot resolve the issue, contact your local Agilent Support representative for more help.

 Tip 
Learn how to effectively troubleshoot your Agilent LC System:
HPLC-0GEN-2201e - Chromatographic Troubleshooting for HPLC e-learning course available from Agilent education
  • Troubleshooting
  • Degassers
  • Pumps
  • Analytical LC
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help