Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
LC Portal
  • Agilent Knowledge Portal
LC Portal
Articles Forced Cold Start Procedure
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • LC Portal Home
  • LC Portal Articles
  • LC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  2 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Analytical LC Systems Preparative LC Systems
  • Component :  LC Column Compartments LC Detectors LC Fraction Collectors LC Injectors and Sample Thermostats LC Pumps LC Valves
  • Product Name :  1100 Series LC System 1120 Compact LC 1200 Series LC System 1220 infinity II LC System 1260 Infinity II Bio-Inert LC System 1260 Infinity II LC System 1260 Infinity II Prime Bio LC System 1260 Infinity LC System 1260 Infinity II Prime LC System 1290 Infinity II Bio LC System 1290 Infinity II LC System 1290 Infinity LC System
  • Task :  Operation Troubleshooting
Related Articles
  • LED Status Indicators

    This Information Applies To: All Agilent LC modules Issue: The LED Status Indicators are common features for all the HPLC modules. The indicators are located on the top right corner of each module, opposite to the power switch led as shown in Figure 1 . They can vary between six different statuses: Grey (led off), Green, Yellow, Red, flashing Yellow or flashing Red. The LED Status is very informative on the module performance and can provide important information for troubleshooting. ...
  • Troubleshooting Remote/Cluster Partner Missing Error

    This Information Applies To: All Agilent LC systems. Issue: A cluster is partnership between modules. Most modules are linked to other modules automatically, without intervention from the user, but sometimes the cluster is configured manually. Examples: 1. To work as a ‘one module’ unit, independent modules must be clustered (e.g.: Two G1361A prep pumps are clustered to make one gradient pump). 2. A valve pressure cluster between the 1290 Infinity pump and external valves allows for correct...
  • How to Update or Downgrade Firmware for Agilent LC and CE Modules

    This Information Applies To: Agilent LC and CE Systems compatible with Lab Advisor Issue: Firmware is updated or downgraded when new features are added, or to fix some system bugs. It also solves compatibility issues when a new module is added to an existing system or a new hardware. Some regulatory requirements can also trigger a firmware upgrade or downgrades. This procedure requires the use of Agilent Lab Advisor Software . Resolution: Platforms and Firmware sets Firmware differs...
  • Where to Find the Firmware Version Info of an HPLC Module

    This Information Applies To: All Agilent HPLC modules. Issue: The firmware is stored on a HPLC module's mainboard. It's a set of parameters, low-level commands, and functions stored on the internal memory of a HPLC module and required for its operation. The firmware consists of two independent components: - a non-instrument-specific part, called Resident Firmware - an instrument-specific part, called Main System Firmware Note: As functions are constantly being optimized, bugs fixed, and compatibility...
  • Troubleshooting AriaMx Temperature Related Errors at the Start Up After Delivery

    This Information Applies To: Agilent AriaMx real-time PCR system Issue AriaMx showing a series of errors after delivery, guide for troubleshooting and resolving temperature-related error. Background When an AriaMx is delivered, it could take few hours for its temperature equilibrate to room temperature. If the instrument is turned on before it has adjusted to room temperature, it could produce a whole series of thermal related errors ( Figure 1 ). Figure 1 : Errors displayed at the...

Forced Cold Start Procedure

Answer

This Information Applies To: All Agilent LC modules (except hosted modules).


Issue:

When a module does not initialize correctly or it is not recognized by the software, it is recommended to call our Remote Support Representatives to troubleshoot the issue. You may be asked to follow the cold start procedure to recover the correct functionality of the module. A forced cold start can be used to bring the module into a defined mode with default parameter settings.

 


Before starting:

Save your methods and data before executing a forced cold start. A forced cold start erases all methods and data stored in the nonvolatile memory of the module, i.e a non saved state.
Exceptions are calibration settings, diagnosis, and repair logbooks, which will not, be erased.

 

Do not use a pencil or a fragile tool to move the switches to avoid breaking it inside the module. We recommend you to use a screwdriver or a pen.

 

Steps to follow:

  1. Switch off the modules.
  2. Disconnect the cables that connect the troubled module to the others.
  3. Take note of the current configuration of the switches that are on the back of the module inside the CONFIG slot (Figure 1). This step is important, because after the procedure you will have to move the switches back to that original configuration. Your module might have 2, 6 or 8 switches depending on the model. All Infinity II Series modules have a 6-switch configuration, except the Hosted modules that use a 2-switch configuration (Multicolumn Thermostat, Flexible Cube, UIB II, and 1290 Infinity Valve Drive). The 8-switch configuration is for 1100 and 1200 Series instruments.
  4. Change the switches configuration according to Table 1.
  5. Switch on the instrument and wait 3 minutes until the initialization is completed.
  6. Switch off the instrument.
  7. Place the switches in their original position.
  8. Reconnect all the cables and switch on all the modules in the stack. 

 

Figure 1. Location of Configuration Switch (CONFIG) and different type of Switches of HPLC modules.

 

 

 

 Table 1: Switches' configuration for a forced cold start. 1 means up, 0 means down.  

2 switches 1 0
6 switches 1 0 0 0 0 1
8 switches 1 1 0 0 0 0 0 1

 

 

Special considerations: 

 

The instrument will always revert to the default state after a cold start. It is especially important for plate types and the metering head size in autosamplers. For example, the 900 µL analytical head upgrade will not be recognized. It must be set again in the software (the same as during initial installation). The sampler will default to the standard 100 µL head.

 

 Tip  Learn more on how to operate your Agilent LC System:
LC user manual library on Agilent.com
 
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Privacy Statement
Terms of Use
Contact Us
Site Help