      • Software Status Indicators 5000 Series ICP-OES
      • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections
      • Checking for a Blocked ICP-OES Nebulizer
      • Cleaning a Blocked Nebulizer in the Agilent ICP-OES System
      • How to Power Cycle Agilent 5000 ICP-OES Systems
      • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
      • Agilent ICP-OES Performance Test
      • Bypassing the AVS for the 5000 Series ICP-OES
      • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS
      • Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Software Status Indicators 5000 Series ICP-OES

Created by Jo Boden Jo Boden 9 months ago | Last modified by Carlos Vargas Carlos Vargas 17 days ago

This Information Applies To:  Agilent 5000 Series ICP-OES (5100/5110/5800/5900)


Issue

The ICP Expert software can help to identify system problems by assessing the status of the instrument components. 

 Tip 
Tip: Hardware LED status indicators are discussed in the article 

Resolution

There are two LED displays that indicate instrument status:

  1. The AVS switching valve indicator light on the bottom right of the main software window shows:
    1. Green: the valve is in the inject position
    2. Yellow: the valve is in the uptake position
    3. Grey: the valve position is unknown.

  2. Status window list accessed through the instrument panel of the ICP Expert software. On the ICP Expert main page, select Instrument menu. 

    The list of status LED color indicates are displayed on the left side of the window. A green LED indicates that the instrument component is ready for use and Table 1 lists the possible causes when the LED is grey, yellow or red.

    Table 1: Status indicators and possible cause

Condition Grey Yellow Red
Plasma Plasma off Igniting or extinguishing Plasma alarm
Optics Polychromator heater temperature rising Polychromator heater fault
Pump Pump fault
Camera Peltier cooling Camera status
Peltier fault
Shutter fault
Water Cooling Water cooling fault
Under flow fault
Over flow fault
Under temperature fault

Plasma Torch Door Plasma torch door interlock tripped
Torch Loader Plasma loader interlock tripped
Pre-optics Beam selector fault
Radial scan fault
Gas Module Gas control fault
RF RF fault
Electronics Electronics error
Switching Valve Switching valve fault
Argon Argon supply low
Argon supply high
IsoMist Communication error

 Tip 
Learn how to effectively troubleshoot your Agilent ICP-OES:
ICPOES-5110-1200e - Agilent 5110 ICP-OES Instrumentation Overview
ICPOES-5900-1200e - Agilent 5800/5900 ICP-OES with ICP Expert: Instrumentation Overview e-learning course available from Agilent education

  • Operation
  • Troubleshooting
  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • ICP-OES Software
  • ICP Expert
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help