This Information Applies To: SPS 4/SPS 3 autosampler, 5000 series (5100/5110/5800/5900) ICP-OES
Issue:
How to change the autosampler probe due to maintenance, blockage or damage.
Steps to follow:
Warning: Always ensure that the power to the autosampler is turned off before installing or removing the sample probe
|
Figure 1. Z-axis description
1. Z-axis Slide 2. Eyelet for tube routing 3. Tube routing to this side 4. Tube Restraint
5. Probe Mount Block 6. Mount Nut 7. Sample Probe 8. Accessories Plate
To Remove the Sample Probe:
Tips: For maintenance purposes, it is recommended to change all sample tubing and connectors when changing the autosampler probe. If the sample probe is being changed due to damage, check the sample tubing and any connectors to ensure they remain in good condition.
|
To Install a new Sample Probe:
Warning: It is important to ensure that the length of the tubing allows the probe arm to move freely in all axes. If it is too tight, it will restrict the probe arm and may cause a movement failure. If it is too loose, the probe arm may become tangled in the tubing as it moves |
Figure 2. Description of probe mount assembly
1. Tube Restraint 2. Probe Mount Block 3. Z-axis Slide 4. Mount Nut
Tips: When using a carbon fiber sample probe, do not run the sample line through the hook (or tube restraint) at the top of the Z-axis slide
|
Caution: Do not use the tube restraint for pre-formed sample probes. It is only used for non pre-formed sample probes |
Figure 3. Sample Line Connectors
1. Cable Tie for tubing through eyelet on bottom of housing 2. Grommet
3. Tube Anchor 4. Colored Markers
Tips: The tube markers on the sample line define the optimal sample length of sample line from the sample probe to the exit point of the autosampler
|
The following lists the color of the tube markers on the sample probe tubes:
Learn how to effectively operate your SPS Autosampler:
SPS 3 User's Guide user manual on Agilent.com |