Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • iLab Operations Software
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Agilent SPS 3 Autosampler Initialization Sequence and Troubleshooting
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  2 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Absorption Systems Atomic Spectroscopy Accessories ICP OES Systems
  • Component :  Atomic Spectroscopy Autosamplers
  • Product Name :  240FS AA 240Z AA 280FS AA 280Z AA 5100 ICP-OES 5110 ICP-OES 55B AA 5800 ICP-OES 5900 ICP-OES 700 Series ICP-OES SPS 3 Autosampler
  • Task :  Operation Troubleshooting
Related Articles
  • Agilent SPS 4 Autosampler Initialization Sequence and Troubleshooting

    Multiple Results
    This Information Applies To: Agilent Spectroscopy Systems, SPS 4 Autosampler Issue: When the Agilent SPS 4 autosampler powers up, it performs an initialization routine, which moves the probe arm to predefined positions on all its axes (X, Y, and Theta...
    • Found In:
    • Articles
    • ,
    • Articles
    • ,
    • Articles
  • Agilent SPS 3, SPS 4 Autosampler Offline Message

    Multiple Results
    This Information Applies To: Agilent Spectroscopy Systems, SPS 3, SPS 4 Autosamplers Issue: The Agilent SPS 3 or SPS 4 autosampler performs its power-up initialization sequence, but displays Sampler Offline status in SpectrAA Software in the Optimize...
    • Found In:
    • Articles
    • ,
    • Articles
    • ,
    • Articles
  • How to Change the SPS Autosampler Probe

    Multiple Results
    This Information Applies To: SPS 4/SPS 3 autosampler, 5000 series (5100/5110/5800/5900) ICP-OES Issue: How to change the autosampler probe due to maintenance, blockage or damage. Steps to follow: Warning : Always ensure that the...
    • Found In:
    • Articles
    • ,
    • Articles
    • ,
    • Articles
  • SPS 4 Alignment and Calibration Wizard

    Multiple Results
    This Information Applies To: Agilent Spectroscopy Systems, SPS 4 Autosampler Issue The sample probe may be displaced from its correct position due to the probe hitting a sample vial or the sides of the wash reservoir. User calibration allows for...
    • Found In:
    • Articles
    • ,
    • Articles
    • ,
    • Articles
  • How to Power Cycle Agilent 5000 ICP-OES Systems

    This Information Applies To: The Agilent 5000 ICP-OES Series and Agilent ICP Expert Software Issue How to reset the Agilent ICP-OES system when experiencing a communication error between the instrument and software, or if the instrument seems to be experiencing issues when trying to light the plasma or run a worksheet. These steps will help to determine if there is a hardware issue with the instrument. Steps to Follow Caution : Save all data in ICP Expert software before shutting...

Agilent SPS 3 Autosampler Initialization Sequence and Troubleshooting

Answer

This Information Applies To: Agilent Spectroscopy Systems, SPS 3 Autosampler


Issue: When the SPS 3 autosampler powers up it performs an initialization sequence that moves the probe arm to predefined positions on the X, Z and theta axes (Figure 1). If the probe arm fails to travel the required distance along any of the axes, the probe will stop moving and LEDs will light to indicate an error. The autosampler cannot be used in this state.


Discussion: 

During the initialization, the probe moves to all the physical limits of the working area of the autosampler (Figure 2) and then returns to the home position. This initialization sequence automatically begins each time the autosampler is switched on. 

 Tip 
Tip: The initialization sequence can run when the main instrument (AA or ICP-OES) is disconnected from the autosampler or is turned off.

Figure 1: Autosampler. 1 Front, 2 Back, 3 Right, 4 Left, 5 Power LED, 6 X-axis, 7 Y-axis, 8 Z-axis, 9 Theta-axis (rotation)

  1. The initialization sequence (Figure 2) is as follows:

    • Probe raises to the extremes of the Z-axis,

    • Probe travels to the extremes of the X-axis,

    • Probe travels to the extremes of the theta-axis,

    • Probe positions itself above the fixed wash reservoir location (final location)

Figure 2: Video showing initialization sequence

Steps to Follow:

If the autosampler fails to initialize:

  1. Turn off the power and move the probe by hand on along each axis and check if there are any physical mechanical obstructions. Correct as required. 

  2. If no physical or mechanical obstructions can be identified, turn the unit back on and allow it to attempt the reset procedure again.

  3. If still failing, the Error LEDs on the autosampler connection panel on the back of the instrument will display a code. There are four LED lights:

  4. The LED pattern, where 1 is on and 0 is off, corresponds to an error code and description (Table 1). Note the error code and contact your Agilent Service Representative.

    LED 1

    LED 2

    LED 3

    LED 4

    Error code

    Error description

    0

    0

    0

    1

    1

    Main printed circuit board (PCB)

    0

    0

    1

    0

    2

    Program memory checksum

    0

    0

    1

    1

    3

    Configuration memory checksum

    0

    1

    0

    0

    4

    RAM test 

    0

    1

    0

    0

    5

    Axis initialization

    0

    1

    1

    0

    6

    X-axis collision

    0

    1

    1

    1

    7

    Theta-axis collision

    1

    0

    0

    0

    8

    Z-axis collision

    1

    0

    1

    1

    11

    X-axis movement

    1

    1

    0

    0

    12

    Theta-axis movement

    1

    1

    0

    1

    13

    Z-axis movement

Table 1: Error codes and description

 Tip 
Tip: If the probe doesn't move at all, and there are no error LEDs on, check the Power LED on the autosampler. If it is off, check if power is being supplied to the autosampler. If it is, there may be another fault that will have to be rectified before the error codes can be utilized.

 

 Tip 
Learn how to effectively troubleshoot your Agilent Spectroscopy System:  Guidelines for Troubleshooting and Maintenance of AA Systems
 

 

 

 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Privacy Statement
Terms of Use
Contact Us
Site Help