This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) and Agilent 700 Series ICP-OES (710/715/720/725/730/735)
Issue
Loss or change of sensitivity in ICP-OES can occur under normal operating practices. This discussion identifies factors that can contribute to sensitivity changes including calibration standards, memory effects, and the sample introduction system. Sample introduction components include:
Calibration standard preparation is a common and frequently overlooked source of sensitivity changes. Check the standard preparation history for:
Prepare new standards if any of the preceding are observed.
Memory effects occur when an element or elements in a single sample are retained in the sample introduction system. These elements bleed out of the system in subsequent sample runs and contribute to increased background. The degree of memory effect depends on the:
Memory effects can be found by examining the analyte background following a sample with a high concentration of the analyte. Memory effects are present if in subsequent samples, the background is highest in the first replicate analysis and reduces in following replicates.
The sample introduction system components can contribute to changes in sensitivity. Agilent recommends routine maintenance to reduce changes in sensitivity. The sample introduction components include:
The autosampler probe flexes during movement into and out of sample tubes. Over time, flexing can cause hairline internal cracks that retain elements and sample contamination. To check the sample probe:
Typically, maintenance and cleaning are performed on other sample introduction components before replacing the probe.
2. Inlet switching valve system
The switching valve system improves sample productivity. Over time, sample deposits can accumulate and block or partially block the valve. To check the switching valve for blockage:
If sensitivity returns to normal levels, clean the valve.
Peristaltic pump tubing delivers consistent sample flow to the nebulizer. It is a consumable item that stretches during normal use and when significantly stretched it delivers variable sample flow. Check tubing daily and look for the following signs of wear:
Tubing has a typical lifetime of one to two weeks based on an eight-hour working day. Replace tubing earlier if these problems are observed.
The nebulizer operates under backpressure to turn the liquid sample into an aerosol. Monitor backpressure through software diagnostics and record the normal operating backpressure for each application. An increase in backpressure can indicate a blocked or partially blocked nebulizer, a decrease in backpressure can indicate a leak. When a change in backpressure occurs visually check the nebulizer for the following:
Replace or clean the nebulizer when these problems are observed.
The sample mist created by the nebulizer expands in the spray chamber and is carried into the sample torch in an Argon stream. Contamination on the inside surface of the spray chamber can cause the mist to bead, resulting in incomplete sample transfer. Check the spray chamber for the following:
Clean the spray chamber when these problems are observed.
The torch innermost tube (injector tube) carries sample to the plasma and it can become blocked. Check for following signs indicate blockage:
Clean the torch acid and aqueous or organic matrices when these problems are observed.
Learn how to effectively troubleshoot and operate your Agilent ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting e-learning course available from Agilent education
|