Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
Articles Preparing the Instrument Calibration Solution for the Agilent ICP-OES System
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • ICP-OES Portal Home
  • ICP-OES Portal Articles
  • ICP-OES Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  ICP OES Systems
  • Product Name :  5100 ICP-OES 5110 ICP-OES 5800 ICP-OES 5900 ICP-OES 700 Series ICP-OES
  • Task :  Maintenance Operation
Related Articles
  • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) and Agilent 700 Series ICP-OES (710/715/720/725/730/735) Issue Loss or change of sensitivity in ICP-OES can occur under normal operating practices. This discussion identifies factors that can contribute to sensitivity changes including calibration standards , memory effects , and the sample introduction system . Sample introduction components include: Autosampler probe Switching valve Pump tubing Nebulizer...
  • 5000 Series ICP-OES Status Indicators

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue Description of ICP-OES 5000 Series Status Indicators (LEDs) The Agilent 5000 Series ICP-OES (5100/5110/5800/5900) has two different LED indicators to advise the status of the instrument. The state of both these indicators determines if the instrument is in a fault condition or ready to operate condition. The ready to operate condition is shown when both indicators are a steady green color. Any other color...
  • Exhaust Requirements for the 5000 Series ICP-OES

    This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900) Issue Correct ventilation is required to ensure proper function of the ICP-OES Resolution The ICP-OES must be located under a flue that is vented by an exhaust fan and ducted to an external vent. Caution : The exhaust system installation must comply with any rules and/or regulations that may be imposed by the local authorities responsible for control of facilities and fixtures in the work place...
  • How to Reset the AVS 6/7 on Agilent ICP-OES Instruments

    This Information Applies To: 5000 series ICP-OES (5100/5110/5800/5900), ICP Expert Issue How to Reset the AVS 6/7 after maintenance or when a red LED is visible Steps to follow Tips : Instructions for changing the AVS switching valve can be found in the ICP Expert Help under “Hardware How to” and “Replace the AVS 4/6/7 Tubing”. Figure 1. ICP Expert Help After replacing the AVS valve and tubing, a red LED is showing on the AVS plate. Figure 2 shows the location...
  • Water Quality Recommendations for ICP-OES Samples

    This Information Applies To: Agilent 5000 series ICP-OES Systems (5100/5110/5800/5900) Issue Discussion of Water Quality for use in the make up of Samples, Blanks, and Standards for ICP-OES Resolution Summary of Recommendations for Water Quality in samples, standards, and blanks for ICP-OES system applications: Low ionic content / resistivity 18 MΩ•cm Low particle content / particles sizes < 0.2 µm Low organic content / Total organic carbon (TOC) Low bacteria content ...

Preparing the Instrument Calibration Solution for the Agilent ICP-OES System

Answer

This Information Applies To: All ICP-OES: 5000 (5100/5110/5800/5900) and 700 (710/715/720/725/730/735) series


Issue

The instrument calibration solution is used when performing the ICP-OES Performance Test and the Wavelength Calibration. The instrument calibration solution can be prepared from 1000 ppm stock solutions (Table 1) according to the calculations described in this article. Alternatively, you can purchase the Agilent ready to use wavelength calibration standard (Table 1).


Instrument calibration solution elements and concentrations

The instrument calibration solution must contain 5 ppm of each of Al, As, Ba, Cd, Co, Cr, Cu, Mn, Mo, Ni, Pb, Se, Sr, Zn, and 50 ppm K in 1% HNO3. Agilent 1000 ppm stock solutions are available as per Table 1 (1 ppm is the same as 1 μg/mL).

Concentration equation:

Use the following equation to calculate the volume of known concentration stock standard required to obtain the desired concentration:

C1V1 = C2V2 

Where:

  • C1 = known concentration stock standard
  • V1 = the volume of the known stock standard required to obtain the desired concentration
  • C2 = desired concentration
  • V2  = final volume

Steps to follow

 Warning 
Warning: HNO3 Is corrosive. Wear appropriate Personal Protective Equipment when handling HNO3
  1. Calculate the volume of each 1000 ppm single element stock standard (Al, As, Ba, Cd, Co, Cr, Cu, Mn, Mo, Ni, Pb, Se, Sr, Zn) required to make up 1 L of 5 ppm working instrument calibration standard:
    1. V1 = C2V2 /C1
    2. V1 = 5 ppm * 1000 mL/1000 ppm
    3. V1 = 5 mL aliquot of 1000 ppm stock standard element is required.
  2. Calculate the volume of the 1000 ppm single element stock standard (K) required to make up 1 L of 50 ppm working instrument calibration standard:
    1. V1 = C2V2 /C1
    2. V1 = 50 ppm * 1000 mL/1000 ppm
    3. V1 = 50 mL aliquot of 1000 ppm stock standard element is required.
  3. Add 5 mL of each of (Al, As, Ba, Cd, Co, Cr, Cu, Mn, Mo, Ni, Pb, Se, Sr, Zn) to a 1 L volumetric flask.
  4. Add 50 mL of K stock standard to the 1 L volumetric flask.
  5. Make up to 1 L with 1% HNO3.
    1. Note: 1%HNO3 is 10 mL conc.HNO3 in 1 L H2O.


Table 1: Agilent Calibration Solutions

Standard description Part number
ICP-OES Wavelength Calibration Solution, Multicomponent, Ready to use 6610030100
Individual stock solutions  
Aluminum (Al) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8242
Arsenic (As) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8246
Barium (Ba) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8248
Cadmium (Cd) standard, 1,000 µg/mL, in 5% HNO3, 500 mL 5190–8328
Cobalt (Co) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8346
Chromium (Cr) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8344
Copper (Cu) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8348
Manganese (Mn) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8483
Molybdenum (Mo) standard, 1,000 µg/mL, in 1% NH4OH, 100 mL 5190–8487
Nickel (Ni) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8491
Lead (Pb) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8475
Selenium (Se) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8519
Strontium (Sr) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8527
Zinc (Zn) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8557
Potassium (K) standard, 1,000 µg/mL, in 5% HNO3, 100 mL 5190–8503

 

 Tip 
Learn how to effectively operate your Agilent ICP-OES:
Agilent 5100 and 5110 ICP-OES  user manual on Agilent.com
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help