This Information Applies To: Agilent ICP-MS Systems
Issue
The Helium cell gas line on the ICP-MS may be contaminated with air and moisture in case of a tank replacement, or if the instrument was not used for an extended amount of time. Purging the Helium cell gas line helps ensure a steady supply of clean Helium to the Cell. It can also reduce the Relative Standard Deviation (RSD) of your tune results. This is especially true in cases where operator observes a high RSD in Helium Mode, compared with low RSD running in No Gas Mode.
Symptoms:
Steps to Follow
Tips: Gas line purge is detailed in the Hardware Maintenance Guide / Appendix D
|
Check that exhaust is working properly and that plasma is turned off.
Figure 1. Dashboard showing plasma status is off.
Figure 2. Dashboard showing instrument is in standby status.
On the hardware pane, click Collision/Reaction Cell > Maintenance
Figure 3. Dashboard showing Maintenance option for Collision Cell.
Click Open Bypass Valve.
Figure 4. The maintenance screen showing the "Open ByPass Valve" was selected.
Increase input gas flow gradually by 0.2 mL/min steps until desired value is reached.
Purge time and flow depend on the total volume of the external gas line, but general set can be 8-10 mL/min for 3 hours or 2 mL/min overnight.
Figure 5. The maintenance screen showing the adjustment of Helium gas flow rate
Note:
Learn how to effectively operate your Agilent ICP-MS system:
ICPMS-MULTI-2200e - Agilent ICP-MS with MassHunter: System Maintenance e-learning course available from Agilent education |