Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Atomic Absorption Portal
  • Agilent Knowledge Portal
Atomic Absorption Portal
Articles AA software - SpectrAA Flags
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • Atomic Absorption Portal Home
  • Atomic Absorption Portal Articles
  • Atomic Absorption Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  8 Feb 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  Atomic Absorption Software
  • Product Name :  SpectrAA Base SpectrAA Pro
  • Task :  Data Analysis Operation Troubleshooting
Related Articles
  • Outlier Setup in MassHunter Quantitative Analysis Software

    This Information Applies To: Agilent MassHunter Quantitative Analysis Software Issue MassHunter Quantitative Analysis allows limits to be set on many result values, and for results outside these limits (outliers) to be flagged in the batch report. The flagged results are easily identified, enabling fast visual identification of outliers, which improves productivity and, reduces the risk of reporting invalid data. The following procedure introduces the use of outliers within the MassHunter...
  • ICP-OES Detector and Instrument Calibration

    This Information Applies To: Agilent 5000 Series ICP-OES (5100, 5110, 5800, 5900) Issue How to calibrate the ICP-OES ICP-OES calibration is a two-step procedure: The detector calibration performs a dark current calibration to correct for detector and electronic background signals. The instrument calibration uses a specified calibration solution to calibrate the wavelengths of the polychromator. ICP-OES wavelength calibration solution is available as either Agilent ICP-OES wavelength...
  • Generating Reports in SpectrAA Software

    This Information Applies To: Agilent SpectrAA Software Issue There are three different ways to generate reports in SpectrAA Software. Generating a Report from Stored Data Automatic Report Generation During a Worksheet Run Automatic Report Generation After a Worksheet Run Requirements Reports may be printed, exported, or generated on screen for viewing. A printer must be set up as default on the PC before printing. Steps to Follow Generating a Report from Stored Data ...
  • RIN Not Calculated or NA for Total RNA Samples in the Agilent Bioanalyzer 2100 Expert Software

    This Information Applies To: 2100 Bioanalyzer, 2100 Expert software, RIN NA, Error 4501, RNA 6000 Nano kit (p/n 5067–1511 ), RNA 6000 Pico kit (p/n 5067–1513 ) Issue: The RIN is not displayed in the software or shows as NA. If the RIN is NA, the sample will be flagged in red and error 4501 will be present with a description such as Unexpected signal... . Steps to follow: Check which assay was selected. A RIN, will only be calculated if a total RNA assay was selected. A RIN will not...
  • Agilent SPS 3, SPS 4 Autosampler Offline Message

    Multiple Results
    This Information Applies To: Agilent Spectroscopy Systems, SPS 3, SPS 4 Autosamplers Issue: The Agilent SPS 3 or SPS 4 autosampler performs its power-up initialization sequence, but displays Sampler Offline status in SpectrAA Software in the Optimize...
    • Found In:
    • Articles
    • ,
    • Articles
    • ,
    • Articles

AA software - SpectrAA Flags

Answer

This Information Applies To: Agilent SpectrAA Software


Issue: 

Explanation of concentration results flags from SpectrAA Software


Under certain conditions, a concentration result may appear with a concentration result 'Flag'. The following table lists each possible concentration result flag and its description. The Priority column indicates the priority number of each flag, Priority 1 = highest priority. If a sample concentration results in more than one error being flagged, the flag with the highest priority is displayed.

The flag may simply be noted for reference purposes or may indicate that corrective action is needed, or one or more standards or samples may need to be rerun.

Flag

Priority

Flag Description

A

6

For Zeeman, the absorbance has exceeded the maximum Zeeman absorbance value set in the Method. Absorbances above the maximum value give erroneous results due to Zeeman calibration rollover.

B

11

The absolute concentration of the solution is less than RDL (Required Detection Limit) and greater than or equal to IDL (Instrument Detection Limit).

D

3

Autodilution error. Volume is less than 1 µL for furnace; factor is greater than 50 for SIPS.

G

4

Low mains voltage.

I

9

Atomic signal absorbance is less than minimum absorbance (user specified limit).

L

13

LCS (Laboratory Control Sample) is out of limits.

M

10

%RSD (percent Relative Standard Deviation) is greater than limit.

N

13

Matrix spike recovery is outside limits.

Q

13

QC (Quality Control) Standard recovery is outside limits.

T

5

Total absorbance is over-range.

U

12

The absolute concentration of the solution is less than IDL.

P

13

Preparation blank absolute concentration is greater than user specified limit. See Note #2.

W

13

QC Spike percent recovery is outside upper and lower limits.

Y

13

The RDL concentration is outside the user specified limits. See Note #2.

Z

13

QC Blank absolute concentration is greater than or equal to the user specified limit.

*

13

Duplicate Relative Percent Difference exceeded.

+

13

Correlation coefficient less than the specified limit.

a

7

Foreground absorbance is over-range. See Flag "a" Note.

e

2

This data item has been edited.

b

8

Background absorbance is over-range (>2 ABS).

m

1

The solution has been measured in read mode.

p

13

PROMT (Precision Optimized Measurement Time) error. This error indicates that the instrument detected a zero or negative transmittance value.

r

13

The solution has been remeasured.

?

13

There are also 10 user definable tests available. The flag for each test is also user-defined. (only available for SpectrAA Pro or CFR (full QCP) software)

 

Tip
Learn how to effectively maintain and operate your Agilent Spectroscopy System:
Agilent 240/280 Series AA Users Guide

 




 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help