This Information Applies To: Agilent ICP-MS systems (7700/7800/7850/7900/8800/8900)
Issue
Drift up is most often a sign of poor cone conditioning.
Drift downwards is often associated with build up on the sample introduction components during an analysis. This build-up is more typical when running higher matrix samples with high percent total dissolved solids (%TDS) which causes salt build-up on the nebulizer, torch injector, cones, and lenses.
Sometimes drift can occur both up and down during the span of an analysis. It is important to evaluate if it is drift, or simply the result of matrix effects from a sample or a problem with poor internal standard mixing. If the drift appears to be non-matrix related, it may be due to either static charge, or an issue in the sample introduction.
Steps to Follow
Figure 1. Gas connections
1. Dilution gas connection, 2. Makeup gas connection, 3. Nebulizer/carrier gas connections, 4. Gas connections on left panel
Figure 2. Connector block
1. Conductive connector, 2. Ground clip on the peri-pump
Figure 3. Plugged connection block
1. Internal standard line plugged
Figure 5. Batch calibration page
1. ISTD column set to <None>
Learn how to effectively troubleshoot your Agilent ICP-MS System:
Agilent 7800/7850/7900 ICP-MS with MassHunter Agilent 8900 Triple Quadrupole ICP-MS with MassHunter e-learning course available from Agilent education |