This Information Applies To: Agilent ICP-MS MassHunter Software 4.5 and later with User Access Control (UAC).
Issue
When trying to access Change User Access Control (UAC) Application and the user gets permissions errors, they are usually related to the configuration of the current user account.
Background
To access Change User Access Control, you must log in to Windows as a user who belongs to the Windows Administrator group, and the user who logs into ICP-MS MassHunter must belong to the OpenLab Shared Services (OLSS) Control Panel ICPMHAdmin group. In addition, User Account Control in Windows Control Panel must be disabled.
If you see the dialog box in Figure 1, your Windows user is not a member of the Administrators group, or the User Account Control Settings in Windows control panel is not set to never notify.
Figure 1. IcpUserAccessCtrl.exe User Account Control dialog box
If you see the error in Figure 2, your OLSS Control Panel user is not a member of the ICPMHAdmin group, or the ICPMHAdmin group has been renamed.
Figure 2. Error: You have no permission to change security settings.
Steps to Follow
Set User Account Control Settings to Never Notify
Tips: For instructions on adding a user to the Administrators group, see the article How to Access Spectroscopy Database Administrator.
|
change user account control settings
(Figure 3; Item 1) and under Best match, click Change User Account Control (Figure 3; Item 2).
Figure 4. User Account Control Settings window
1. Slider set to Never notify, 2. OK button
Check if a User is a Member of the ICPMHAdmin Group
Tip: Contact your system administrator if you do not have Group membership: ICPMHAdmin. They will either need to add you to the group, or a different user who is already a member of the ICPMHAdmin group will need access Change User Access Control. Proceed to step 3 for instructions on how to edit a user.
|
Confirmation: Verify that the group name is exactly ICPMHAdmin. Any change to the group name will disrupt access to Change User Access Control. If the group name has changed, you must edit the group and rename back to ICPMHAdmin.
|
Figure 7. My Settings - OLSS Control Panel
1. My Settings in left panel, 2. Group membership: ICPMHAdmin
Edit User Group Membership in OLSS Control Panel
Figure 10. My Settings - OLSS Control Panel
1. My Settings, 2. Role membership: Everything
Figure 11. Users - OLSS Control Panel
1. Users Tab 2. Admin User Selected 3. Edit User Button
Figure 12. OLSS Control Panel - Edit User
1. Group Membership tab 2. ICPMHAdmin Box Checked
To launch the Change User Access Control application, from the Windows Start button, click the ICP-MS MassHunter Workstation folder > Change User Access Control (Figure 13).
Figure 13. Launching Change User Access Control
1. Start Button 2. ICP-MS MassHunter Workstation (folder) 3. Change User Access Control Application