Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC/MS
    • LC
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles How to Install a Capillary Column on 5977 Series GC/MSD
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  2 Aug 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC/MS Systems GC Systems
  • Component :  GC/MS Single Quadrupole
  • Product Name :  5977 Series GC/MSD
  • Task :  Maintenance Operation
Related Articles
  • Using a Column Installation Tool to Install a Capillary Column on Agilent 5977A/B GC/MSD

    Multiple Results
    This Information Applies To: Agilent GC/MS 5977A, 5977B Content This article shows how to install a capillary column on 5977A/B GC/MS using a column installation tool. Steps to follow: Warning : – The GC and MS operate at high temperatures...
    • Found In:
    • Articles
    • ,
    • Articles
  • How to Replace the Electron Multiplier Horn - 5977 Series GC/MSD

    This Information Applies To: Agilent 5977A/B/C GC/MSD Issue The Electron Multiplier Horn must be replaced when it is contaminated or the EMV value reaches 3000 V with a clean ion source. Items Required Electron multiplier horn (5977B/C: p/n G7002-80103 , 5977A: p/n G3170-80103 ) Gloves, clean, lint-free Large (p/n 8650-0030 ) Small (p/n 8650-0029 ) Long nose pliers or tweezers Steps to Follow Warning : The analyzer, GC/MS interface, and other components in the analyzer...
  • How To Pump Down (Start) the GC/MSD - EI - MassHunter

    This Information Applies To: Agilent 5977A/B GC/MSD with MassHunter GC/MS Acquisition Software Issue: This article describes the procedure to safely pump down (start) 5977A/B GC/MSD with Electron Ionization (EI) Steps To follow: Before you turn on the MSD: The vent valve should be open slightly. All other vacuum seals and fittings must be in place and fastened correctly. The front side plate screw should not be tightened, unless hazardous carrier, JetClean system hydrogen, or reagent...
  • How To Vent (Shut Down) the GC/MSD - EI - MassHunter

    This Information Applies To: Agilent 5977A/B GC/MSD with MassHunter GC/MS Acquisition Software Issue: This article describes the procedure to safely vent (shut down) the 5977A/B GC/MSD with Electron Ionization (EI) The MSD will be damaged by incorrect venting. A diffusion pump will push back upstream vaporized pump fluid onto the analyzer if the MSD is vented before the diffusion pump has fully cooled. The turbo pump will be damaged if it is vented while spinning at more than 50% of its normal...
  • How to Remove and Reinstall 5977 GC/MSD Ion Sources (EI/CI/HES)

    This Information Applies To: Agilent 5977 GC/MSD Issue GC/MSD maintenance or repair tasks will require removal and reinstallation of the ion source within the analyzer assembly. Filament exchange Replacement of insulators Ion source cleaning. Types of Sources It is important to identify the ion source type correctly when removing and re-installing an ion source to ensure that the electrical leads are reconnected correctly. Ion source type label is on the heater/sensor part....

How to Install a Capillary Column on 5977 Series GC/MSD

Answer

This Information Applies To: 5977A and 5977B  


Content: This procedure is to guide the installation of a capillary column on 5977A/B GC/MS with a standard column nut or self-tightening nut.


Warning Warning: 
a. The GC and MS operate at high temperatures. Do not touch any parts of GC and MS until you are sure they are cool.
b. Always wear safety glasses when handling capillary columns. Use care to avoid puncturing your skin with the end of the column.

 

 Tip Tip: Always wear clean gloves to prevent contamination while handling any parts that go inside the GC or the analyzer chambers.
  1. Vent the MS, see How To Vent (shut down) the GC/MSD - EI - MassHunter. Open the vent valve by turning the knob counterclockwise.

  2. Install a capillary column in the inlet, see Capillary Column Installation Quick Reference Guide. Alternatively, use a Self Tightening column nut (part number G3440-81011) with short graphite/vespel (15%/85%) ferrules to install the column.

  3. Condition the column.

  4. Identify the ion source:

    1. 5977A/B SST EI, Inert EI, Xtr EI, CI ion sources are identified by opening the door and visually examining the source.

    2. 5977B High Efficiency Source (HES) ion source is identified on the front cover.

Installing the column in the MS:

  1. Open the analyzer chamber and ensure that the end of the GC/MS interface is visible:

    1. The 5977A/B SST and inert EI interfaces will have the end of the column will be visible inside the analyzer chamber: 

    2. The 5977A Xtr EI, 5977B Xtr / HES EI and 5977A/B CI. Remove the tip seal assembly (including ceramic tip seal, spring, and tip cap) to make the tip of the column visible:

      5977A Xtr EI 1-Tip seal 2-Tip cap 5977B Xtr and HES EI 1-Tip seal 2- Tip cap 5977A/B CI 1-Tip seal and spring
       
  2. Slide the column into the GC/MS interface. The tapered end of the ferrule must point towards the nut. Alternatively, use a self tightening nut (part number G3440-81011) for MS interface. A self tightening nut is also available for the GC inlet (part number G3440-81011).

      
    1- GC/MS interface 2- Ferrule 3- Nut 1-Self tightening nut for inlet, 2-Self tightening nut for MS
  3. Adjust the column by pushing the column through the ferrule, cut the column, then pull back until it extends the following distance from GC/MS interface:

    1. SST/Inert/Xtr EI source, 1 to 2 mm.

    2. High efficiency source, 4 to 5 mm. 

    3. CI Source, about 1 mm.

       Tip Tip: Use fused silica tubing cutter to cut the column. The cut end of the column must have be flat with no cracks or jagged edges. Use the flashlight and magnifying loupe if necessary to see the end of the column inside the analyzer chamber. Do not use your finger to feel for the column end.
  4. Tighten the nut:  

    1. Self-tightening nut: Turn the nut in the clockwise direction, continue to tighten until you feel the ferrule grip the column. 

    2. Standard nut: Finger tighten the nut then use a spanner to tighten the nut 1/4 to 1/2 turn. Check the nut’s tightness after one or two heat cycles and tighten as appropriate.

       Caution 
      Caution: Over-tightening the nut may damage the column or block the column flow.
       
  5. Check the GC oven to be sure that the column does not touch the oven walls.

  6. Tighten the nut 1/4 to 1/2 turn.

  7. Check the nut’s tightness after one or two heat cycles as it may loosen. Tighten if necessary.

  8. For 5977A Xtr EI, 5977B Xtr and HES EI and 5977A/B CI interfaces:

    1. Install the tip seal spring

    2. Align and gently slide the ceramic tip seal, next step

    3. For 5977A with Xtr source, slide the tip cap and tighten two screws onto the tip cap.

    4. For 5977B with Xtr and HES source, screw the knurled nut onto the end of the interface.



      5977A with Xtr source 1-Spring 2-Tip seal 3-Tip cap GC/MS CI source1-Tip seal and spring 5977B with Xtr and HES source 1-Knurled nut 2-Spring
  9. Check the alignment of the ion source and the interface tip seal for CI GC/MS and EI GC/MS with an Xtr and HES source. When the ion source is aligned correctly, the front analyzer chamber can be closed with the interface tip seal. spring tension the only applied pressure. 

  10. Close the analyzer chamber door.

 Tip
Learn more on how to effectively operate your GC/MS:
Agilent 5977 GC/MSD with MassHunter Workstation e-Learning course available from Agilent education. 
 
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel
Contact Agilent Support

Still Need Assistance?

Contact Agilent Support >

Privacy Statement
Terms of Use
Contact Us
Site Help