This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer
Issue
After starting a run, the run is aborted with the following error: Current = 0.0uA, Expected range: 5 - 3000.
Steps to Follow
Check...
This Information Applies To: Parallel capillary electrophoresis platforms, 5200/5300/5400 Fragment Analyzer, ZAG DNA Analyzer, Femto Pulse, Oligo Pro II
Issue:
Disposing used capillary arrays
Resolution:
Arrays can be discarded in regular...
This Information Applies To: 5200, 5300 and 5400 Fragment Analyzer, ProSize software, DNF-471 RNA kit, DNF-472 HS RNA kit
Issue:
The DV 200 score is a quality score for evaluating quality of RNA derived from formalin-fixed paraffin-embedded (FFPE...
This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer
Issue:
Dirt or dust on the capillary array window can cause baseline shifts or elevation of the baseline. Figure 1 shows an example electropherogram...
This Information Applies To: 4200, 4150 and 2200 TapeStation Systems
Issue
The expiration date displayed in the Agilent TapeStation controller software is different to the one printed on the ScreenTape device.
Background
The expiration date...