This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, Oligo Pro II, and ProSize data analysis software
Issue
After the PDF report has been generated in ProSize, the report file is saved with a size of 0 bytes and will not open.
Background
A PDF report can be generated from the ProSize software by clicking File > Generate Report. In the Generate PDF dialog box that appears, you can select what you would like to have included in the report and where you would like the PDF to be saved. The PDF will be generated after you click Generate PDF Report.
Resolution
If the listed advice does not help, try uninstalling and reinstalling the ProSize analysis software.
As a temporary workaround, try using a different PC with ProSize installed to export the PDF and report the issue to the local Agilent support representative.
Learn how to effectively maintain and operate your Agilent parallel capillary electrophoresis systems:
5200, 5300 and 5400 Fragment Analyzer User Manual on Agilent.com
Femto Pulse User Manual on Agilent.com
ZAG DNA Analyzer User Manual on Agilent.com
Oligo Pro II User Manual on Agilent.com
Prosize Data Analysis Software User Manual on Agilent.com
|