This Information Applies To: Agilent Flame Atomic Absorption (AA) Spectrometers (50/55/240/280 AA)
Issue:
How to optimize the Agilent Flame AA for better instrument performance and response stability
Steps to follow:
Use Figure 1 as reference for the procedure on the Flame AA optimization...
This Information Applies To: Agilent Spectroscopy Systems, SPS 4 Autosampler
Issue
The sample probe may be displaced from its correct position due to the probe hitting a sample vial or the sides of the wash reservoir. User calibration allows for minor adjustments to center a misaligned probe using...
This Information Applies To: Agilent SpectrAA Software
Issue
There are three different ways to generate reports in SpectrAA Software.
Generating a Report from Stored Data
Automatic Report Generation During a Worksheet Run
Automatic Report Generation After a Worksheet Run
Requirements...
This Information Applies To: Agilent Flame AA Systems: 55B, 240FS, 280FS
Issue:
After cleaning and re-installing the flame atomic absorption system components, it is important to optimize the burner position, nebulizer uptake rate, flame conditions, glass impact bead , and mixing paddles . Proper...
This Information Applies To: All Agilent Atomic Absorption Spectrometry Systems (AA50, AA55, AA240, AA280)
Issue: This article describes when and how to clean an AA burner.
Background
For optimum and safe performance, burners (see Figure 1 ) should be cleaned after each day of use. Particularly...