Wiki Varian ion trap 3800 GC 4000 MS RF Ramp Adjustment tips
  • Files
  • Wiki
  • More
  • Cancel
  • New
User Contributions requires membership for participation - click to join
  • Welcome to the User Contributions Wiki Page!
  • Fail to load method on MassHunter Acquisition B 06.01
  • Hi-vacuum gauge turning off intermittently/error code:[69][133][159][223]
  • Operating MassHunter and OpenLab in the same computer (NOT A RECOMMENDED USE)
  • Printing out your MassHunter Acquisition method
  • Recovering your .m MassHunter Acquisition analysis method from .d data file
  • Turbo pump failure: Agilent MSD 5977A series with Agilent turbopump
  • Varian ion trap 3800 GC 4000 MS Air/Water check tips
  • Varian ion trap 3800 GC 4000 MS Calibration Gas Adjustment tips
  • Varian ion trap 3800 GC 4000 MS Ion Lens Tuning (131) tips
  • Varian ion trap 3800 GC 4000 MS RF Ramp Adjustment tips

Varian ion trap 3800 GC 4000 MS RF Ramp Adjustment tips

Created by yamame yamame over 2 years ago | Last modified by Agilent Agilent over 1 year ago

Manual: Varian GC 3800, Varian 4000 MS

We had a Trap Frequency Calibration failure and the service team helped us to fix it.

The service team operated the following procedure:

Air/Water check

Ion Lens Tuning (131)

Calibration Gas Adjustment

RF Ramp Adjustment

 

Here I will share about RF Ramp Adjustment for Varian 4000 MS.

I will share tips that I learned from the service team that I found it hard to find in the manual. 

 

Trap Frequency Calibration failure

 

Do manual tune: RF Ramp Adjustment

see manual P. 42 Varian 4000 MS

 

 

1) Load service method (if you have, if you don't, skip this but do not save/overwrite your analysis method), and go to manual control tab

2) Turn on the trap by clicking the image of the trap or checking the box. When the trap is on, words in the image of trap will turn green.

 

3) Choose Active method segment Number 2 : background /HMN. You can change active method segment by up and down arrow.

4) Chose RF Ramp adjustment from Checks and Adjustments.

 

5) Hit Start.

 

6) Check if the line is linear.
If not, stop tuning and call the service.

 

 

 

7) In Adjustment Result, you want to make this number and gauge as low as you can (number around 550-600). You can adjust this number and gauge by tuning the RF Adjustment screw head (RF adjustment screw head is inside the MS front door. Door will be open by just pulling to right.).

8) By tuning the screw head you will find the lowest point, if you go further than that, the number will go higher again.
This is a coil adjustment. You want to try both directions to make adjustments.

9) Click Done when you finish.

 

 

Baking out the system: see manual P. 50 Varian 4000 MS

 

 

This is a pretty old system and you might not use them anymore, but if you do, I hope these tips help!