Wiki Varian ion trap 3800 GC 4000 MS Ion Lens Tuning (131) tips
  • Files
  • Wiki
  • More
  • Cancel
  • New
User Contributions requires membership for participation - click to join
  • Welcome to the User Contributions Wiki Page!
  • Fail to load method on MassHunter Acquisition B 06.01
  • Hi-vacuum gauge turning off intermittently/error code:[69][133][159][223]
  • Operating MassHunter and OpenLab in the same computer (NOT A RECOMMENDED USE)
  • Printing out your MassHunter Acquisition method
  • Recovering your .m MassHunter Acquisition analysis method from .d data file
  • Turbo pump failure: Agilent MSD 5977A series with Agilent turbopump
  • Varian ion trap 3800 GC 4000 MS Air/Water check tips
  • Varian ion trap 3800 GC 4000 MS Calibration Gas Adjustment tips
  • Varian ion trap 3800 GC 4000 MS Ion Lens Tuning (131) tips
  • Varian ion trap 3800 GC 4000 MS RF Ramp Adjustment tips

Varian ion trap 3800 GC 4000 MS Ion Lens Tuning (131) tips

Created by yamame yamame over 1 year ago | Last modified by Agilent Agilent over 1 year ago

Manual: Varian GC 3800, Varian 4000 MS

 

We had a Trap Frequency Calibration failure and the service team helped us to fix it.

The service team operated the following procedure:

Air/Water check

Ion Lens Tuning (131)

Calibration Gas Adjustment

RF Ramp Adjustment

 

Here I will share aboutIon Lens Tuning (131) for Varian 4000 MS.

I will share tips that I learned from the service team that I found it hard to find in the manual. 

. 

 

Trap Frequency Calibration failure

 

Do manual tune:  Ion Lens Tuning (131)

see manual P. 49 Varian 4000 MS

 

 

1) Load service method (if you have, if you don't, skip this but do not save/overwrite your analysis method), and go to manual control tab

2) Turn on the trap by clicking the image of the trap or checking the box. When the trap is on, words in the image of trap will turn green.

 

3) Make a check in Calibration Gas box.

 

4) Choose Active method segment Number 4 : 131
You can change active method segment by up and down arrow.

 

5) Check if you see 131 m/z.

 

6) Uncheck the calibration gas box.

 

7) Check if the 131 m/z spectra go down.
If it does not go down, it means you had a Cal gas contamination in the MS. Do bake out.

 

8) Turn off the trap by unchecking the box or click the image.

 

Baking out the system: see manual P. 50 Varian 4000 MS

 

 

 

This is a pretty old system and you might not use them anymore, but if you do, I hope these tips help!