Problem in LCMS G6125B

Hello, I have a strange peaks were appear at every injection, even when I tried to inject ACN only , I got high response unknown peaks, so I tried to clean ESI spray zone with methanol, but no thing was changed, so How to resolve this issue?

So, Any suggestions please...

Parents Reply
  • When you talk about peaks, do you mean peaks in the chromatogram or background ions in all of the mass spectra.

    • If they're background ions in the mass spectra, what are the masses that you see?
    • If they're additional peaks in the chromatogram, are you running a gradient separation? 
    • Do the peaks appear if you do a run without making an injection (run a sample without specifying a vial location, and the system will just start a run.)?
    • Do the peaks change in area with injection volume?
    • Do the peaks change in area, depending on the length of time between injections?

    Sorry for all of the questions!

    /Andy