Problem in LCMS G6125B

Hello, I have a strange peaks were appear at every injection, even when I tried to inject ACN only , I got high response unknown peaks, so I tried to clean ESI spray zone with methanol, but no thing was changed, so How to resolve this issue?

So, Any suggestions please...

Parents Reply Children
  • Hello, If the impurities appear as peaks with retention time, then the impurities cannot come from the mass spectrometer, i.e. the source, but from the LC with all possibilities such as contaminated fluid (water or organic), dirty autosampler from previous samples, etc.

    Above possibleties, i already mentioned in starting of my replies to check hplc module/columns via bypassing with union to confirm that problem is with ms or hplc..