We have agilent 6545 qtofms. The quad tune in neg mode is failing due to failed wide peak width for 1633 ion. any suggestions to troubleshoot. please see the image. Thanks
Find maintenance, best practice, and troubleshooting articles.
Hi
I was wondering if Dorths advice helped you?
Hello,
Yes the suggestion by Dorth has helped. We have done a system tune and went back to check tune.
We had the same problem another time and we are also at that time where we have to change the glass capillary and it was fine.
Harsha
Hello,
Yes the suggestion by Dorth has helped. We have done a system tune and went back to check tune.
We had the same problem another time and we are also at that time where we have to change the glass capillary and it was fine.
Harsha
Thanks for getting back and closing the loop.
Mitch